randy » Thu Feb 19, 2009 9:28 pm
I have come upon a problem that has developed with our 5975C and cryogenic preconcentrator analyzing 6L SUMMA canisters. I say developed because the MS seemed to chug along just fine for its first 3 months of use. Now when I inject samples that contain water, the response of the MS decreases significantly. If I go back to injecting internal standard without anything else, I can watch the peak areas of the ISTDs go back up. The preconcentrator does have a form of water control, but I don't think I can eliminate the water completely from these samples.
If the 5975 is really this sensitive to water, how does anyone successfully analyze purge and trap or canister samples with this model? I know our 5971 isn't this sensitive to water. Our 5973 is more sensitive to water than the 5971, but I don't think it's as sensitive to water as the 5975.
Is there anything I can tweak with the MS to lessen the effects of water? I don't know how much is being put onto the column, but it does raise the baseline for about three minutes.