Retention time shortened abruptly

Posted: Wed Feb 18, 2009 2:27 am
by yinkheay
Hi. I'm pretty new in HPLC. I'm assigned to separate 2 DNA fragments (300bp and 2500bp) using HPLC.

Recently I encounter a problem whereby my retention time shortens by about 7 minutes. Not only that, initially there were 2 peaks coming out (at 24 and 29 min) but, with the same sample, same gradient, pressure and flow rate, now the peaks overlapped and come out at 17.5 min.

What could be the reason for this abrupt change in retention time? What should I do to separate the peaks?

Posted: Wed Feb 18, 2009 3:18 am
by Uwe Neue
First things to check: does the pump work properly? Have I made the solvents correctly?

Posted: Wed Feb 18, 2009 6:46 am
by yinkheay
Flow rate and pressure are fine. And the solvents are freshly prepared and are the same with that used in previous successful runs.

Posted: Wed Feb 18, 2009 1:51 pm
by Uwe Neue
Next questions: what are the solvents? I assume you are running a gradient, please describe... What's the temperature? Etc....

Posted: Thu Feb 19, 2009 1:07 am
by yinkheay
Column: Waters Gen-Pak FAX (4.6 x 100mm) (an anion exchange column)
Buffer A: 25 mM Tris-HCl, 1 mM EDTA, pH 8.0
Buffer B: 25 mM Tris-HCl, 1 mM EDTA, 1.0 M NaCl, pH 8.0
Flow rate: 0.5 ml/min
Gradient: Load at 100% A. Then 0% to 50 % Buffer B in 1 minute. Then 50% to 60% Buffer B in 30 minutes.


I don't have a temperature controller. But, because my lab ambient temperature only range from 30-35 celcius. I believe this will not affect the HPLC much.