Troubles with analysis of acetonitrile with GC-FID
Posted: Fri Nov 22, 2019 2:48 pm
by Sfverhae
Hello
I have some questions about my method for testing acetonitrile in an aqueous sample. I use a GC-FID
First my GC settings (GC 6890N):
inlet temperature: 200 °C
Detector temperature: 250 °C
Splitless injection: 1 microliter was injected with autosampler
Temperature program:
0-3 min: 60°C
after that it rise with 80°C/min to 220 °C then it holds that temperature for 2min.
Hydrogen and nitrogen: 40 mL/min
Air: 400 mL/min
Column: HP FFAP 30 m * 320 μm * 0.25 μm
1) I don't get consistent results. I've read that it's hard to get good results with a GC-FID for testing acetonitrile. But it's not impossible I hope? Because I don't really have an alternative. I found another method with GC-NPD but the upper range of this method is 5 mg/L and my range of concentration lies between 5 mg/L and 250 mg/L so that's also not an option.
2) maybe my column is broken. On the picture you see that at the end there're a lot of peaks and the baseline gets higher. I searched this and maybe this is columnbleed? Can I fix this or will the results be better with a new column?
Thanks in advance
Re: Troubles with analysis of acetonitrile with GC-FID
Posted: Fri Nov 22, 2019 2:53 pm
by Sfverhae
Here's a link to the picture I'm talking about.
https://imgur.com/a/UINKxvP
Posted: Fri Nov 22, 2019 4:51 pm
by James_Ball
It may be the FFAP column. Reading about it, the stationary phase contains acid, and acetonitrile can be broken down by acid. Maybe you need a column such as a normal Wax column or even one of the 624 phase columns https://www.agilent.com/en/product/gc-c ... cb-columns
The other option is to use either Headspace sampling or Purge and Trap, since acetonitrile is volatile enough to use those methods.
Posted: Fri Nov 22, 2019 10:39 pm
by Consumer Products Guy
USP assay for ethanol uses ACN as the internal standard (instead of the more-commonly utilized n-propyl alcohol) and FC-FID, so my supervisor wanted us to do and validate our procedure using ACN. We used water as solvent, and didn't have any reproducibility issues.
No - we did not use FFAP.
Posted: Sat Nov 23, 2019 4:20 am
by Sfverhae
Which column did you use then?
Posted: Sat Nov 23, 2019 4:23 am
by Sfverhae
Thanks for the answer. The column that I use is made for analyzing organics acids. But if I inject acetic acid (400 ppm) in a water sample than I don't get any peak. Only the peaks at the end like I always have. So that's not normal I guess?