Agilent G1322A degasser - strange failure mode
Posted: Wed Nov 20, 2019 1:47 pm
by ChristofferBR
Hi all,
I am troubleshooting an Agilent 1100 HPLC system which shuts down completely during use. This was traced to be caused by a 'fault' signal sent by the degasser to the pump. Unplugging the signal cable from the degasser yields a functioning, yet degasser-less HPLC.
Troubleshooting the degasser, coupling the pressure sensor to the pump and solenoid, I see pattern where it pumps down, then within 20 seconds releases pressure through the solenoid, then pumps down again, repeat.
Vacuum obtained on the vacuum output terminals gives a value slightly above 600 mv.
When installed in the hplc system, the symptoms are completely classic "try making vacuum for 8 minutes, if not then fault/red light" situation.
Has anyone had any luck solving this?
Thanks in advance.
Christoffer
Re: Agilent G1322A degasser - strange failure mode
Posted: Wed Nov 20, 2019 2:39 pm
by Consumer Products Guy
I've replaced the vacuum tubing in these because of cracks/leaks.
I have even used vacuum tubing from the auto parts store !
Posted: Thu Nov 21, 2019 2:48 am
by alopes
Hi Christoffer,
I can help you with this.
As you know, there are 3 serviceable components to the degasser: vacuum chamber, solenoid valve, and pump. The pressure sensor is coupled directly the power supply and is generally replaced as an entire unit. I should also mention that it is unlikely that there is an issue with your sensor / electronics as, in my experience, these rarely fail.
The first thing I would recommend doing is replacing the tubing on the unit. The G1322A systems, depending on the year they were manufactured or refurbished, will probably have black polyethylene tubing with a metal spring on the inside. The tubing dries out over the years, especially at the points where they are connected, which causes cracks and leaks in the system. I would highly recommend purchasing new tubing from Agilent (PN: G1379-67310 - this works in G1322A systems) or from your local hardware store.
I would then begin isolating each component. Once you power up your degasser, the solenoid valve will switch off and pull vacuum from the top (you can feel it with your finger). Then after ~15 seconds, you will hear a click in the solenoid valve and there will no longer be suction from the top of the valve but rather diverted to the vacuum chamber. Your pump will pull vacuum until it reaches below a certain threshold and the indicator light will turn off (will go from yellow to off). If the pressure is too high, the yellow light will remain on. As a quick test, I would recommend connecting a 20 mL syringe to your pressure sensor (the small chip directly on the back of the vacuum chamber with a tube connected to it) and to pull vacuum on it. If the light turns off when you pull vacuum using the syringe, you know your sensor is good.
I would then go look at your pump. The pumps should be serviced every 3-5 years since they use two rubber flaps (one on each side of the pump) to drive the suction process. It was recommended that these rubber flaps are replaced every 3-5 years to maintain optimal performance, but end users only service degassers only when something goes wrong. You can purchase replacement pumps (Iwaki ERC-6210 or APN-4002) from either Agilent or Ebay to see if that helps. You may also want to carefully inspect your solenoid valve to make sure it isn't leaking - there are often micro leaks which cause issues.
It is possible there is a leak in the vacuum chamber; however, this usually only occurs when using solvents such as chloroform, DMF, THF, or other strong organic solvents that can dissolve glue / epoxy. To test the functionality of this, I generally hook this up to my building vacuum system with the pressure sensor, pull enough vacuum to turn the yellow light off (wait another 3-4 minutes after this), turn the vacuum off, and see if the system remains under pressure for at least 2-3 hours. If this passes, then it is likely your solenoid valve, tubing, or pump at fault - it could even be all 3. When doing this test, make sure all of your lines (A,B,C,D) are filled with solvent (water is preferable due to low compressability).
Feel free to contact me if you require any help with this: Lopesa (at) mit (dot) edu.
Best,
Aaron
Posted: Wed Nov 27, 2019 12:03 am
by Multidimensional
"caused by a 'fault' signal sent by the degasser to the pump" - - That is correct and means you have your HPLC degasser module correctly connected and installed. The remote cable connection to the degasser was put there as a last ditched effort for the system to tell the customer that their degasser module has been broken for some time and requires professional service. *The RED error status light was already on. Long before the RED light came on, the YELLOW status light on the degasser was lit up all of the time indicating the degasser required service. These features are in place to let you know there is a problem which needs to be addressed.
"coupling the pressure sensor to the pump and solenoid, I see pattern where it pumps down, then within 20 seconds releases pressure through the solenoid, then pumps down again, repeat.
." - - For the G1322A, with just the pump connected to the sensor, it should pull a very strong vacuum, hit the set point and the pump should turn off for a LONG time. If you hook up the solenoid valve between the sensor and vac pump, then the pump should do the same thing, but this time with the solenoid stopping the pump a bit earlier in the cycle to vent the pump (but still hold vac). The slightly greater than 600 mV signal is typical, but because most of the circuits loose calibration over time it no longer is reliable of any actual vacuum reading so of little value in troubleshooting exc to show you have vacuum.
If your vac pump is bad, then so is your vacuum valve too. The valve acts as a sort of prefilter for the pump (not by design, but by its position inline). As the parts break down, the valve sends contamination to the vac pump. A weak valve damages and accelerates the wear on the vac pump. Replace them both, and also, all of the internal vacuum tubing too. These are normal wear-tear items which often last no more than 5 years.
Please, NEVER EVER install "Auto-Store" tubing in this or any other HPLC degasser module. Very bad advice. DO not use the G1379-# tubing kit either (that is the wrong kit for this module). Only ultra high purity vacuum rated tubing must be used in this application. If you use cheap tubing from the auto or home center store (usually it is windshield washer fluid tubing), then your degasser will now act as a contamination source for your entire HPLC (or worse yet, an LC-MS system). All of those plasticizes inside that cheap tubing will be mixed with the mobile phase vapors/condensate and re-introduced back into your mobile phase and column. We have serviced many degasser modules over the years where some amateur did just that and contaminated a company's instrument. Use the correct parts for the job.
You did not mention if you checked the condition of the G1322A's 4-CH vacuum chamber (the most expensive part in the whole module, ping). Replacing the vac pump, valve and/or tubing in a system with a leaking or contaminated 4-Channel vac chamber will lead to early failure of the vac pump and vac valve, wasting money and time. If possible, have the vac chamber professionally tested for leakage rate (which requires professional vac equipment and someone in industry experienced in doing this test that understands what the readings mean and what is pass/fail for a leakage rate) and checked for contamination, esp if the degasser is more than 5 years old. If these services are not available locally and you suspect the chamber is leaking, it may be cheaper to buy a new degasser rather than wast money swapping parts.
For service of the G1322A in Europe, Agilent Techn is probably your best bet. In the USA, for the lowest cost, professional service of any model or brand of HPLC degasser see this info for the repair services we offer; http://www.chiralizer.com/uploads/8/9/7 ... s_page.pdf