Peak in Blank Injection-RS

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Hi guys,

In the blank(diluent) injection,for RS method of a NCE molecule, am getting a peak exactly at a RT(about 24 min.) of main peak.
Diluent is ACN:Buffer=1:1,
Buffer here is CH3COONH4+dil. CH3COOH.
I have tried no. of expts to identify its origin.
viz. variations in system,column,chemicals but is not successful.
No matter what is done , this peak is somehow not affected and it appears same in all cases.(area is almost same )
Column is YMC Pack pro C-18(150*3 mm,5 µm),
UV is 230 nm
flow rate is 0.5 ml/min for the gradient elution

Time A B
0 70 30
27 10 90
35 10 90
37 70 30
45 70 30

MP(A)- BUFFER
MP(B)- ACN : MEOH+dil. CH3COOH= 600:400+2.5 ml

Any thoughts will be highly appreciated...![/quote]
Shrikant.

"variations in system,column,chemicals but is not successful. "
so,I am confused. Try to change diluent (ACN:Buffer=1:1)
to other rate (ACN:Buffer=1:2,1:3 etc) to see whether there will be any variation of blank peak, thus may be find the origin.

Shrikant,

Try to search the forum for contamination of the mobile phase A (contamination of the water used in the mobile phase). You'll find several threads about diagnosting ghost peaks due to the present contamination in the HPLC water.

Hope this helps

This one comes up often enough that I've posted a brief (about 15 minutes) mini-seminar on our web site:
http://www.lcresources.com/more_resourc ... hp?f=3&t=5
You will have to register (free) and log in to access it.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
4 posts Page 1 of 1

Who is online

In total there are 4 users online :: 3 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot], Google [Bot], Google Adsense [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry