Port for mix of sample in Agilent 1200 using Empower

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
How to set wash port for mix of sample in Agilent 1200 using empower?
I'm trying to set up pretreatment program at Autosampler for my test in Agilent 1200 using Empower 3. As per my method, In Function section, I have to set Mix in Wash Port. I couldn't able to find this option in program.
How to set wash port for mix of sample in my program? Is it specific program? Or Do I need to setup a new vial for mixing process?

Thanks in Advance,
Satheesh
I made this a new topic so that someone might respond...
Thanks,
DR
Image
Not the "wash port". No washing needed with any model of an Agilent 1200-series A/S. Just use program injector time-table menu in Empower (or any other CDS) to program the steps you need in the A/I.
3 posts Page 1 of 1

Who is online

In total there are 25 users online :: 0 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 25 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry