i have worked with epa method 8270 (semivolatile organics by gcms for about 10 years, but only in the past couple of years have i had trouble determining method detection limits for these 2 compounds.
i am spiking 5 µg, and concentrating the extract to 1 mL, which translates to 5 ng on-column. recovery of 4-chloroaniline ~10%, aniline even worse.
a little experimentation seems to have isolated the problem to the concentration step (k-d concentrators on steam bath), and the presence of the acetone in the solvent.
the mdl spike mix contains about 75 compounds, none of which has presented a problem before, other than benzidine, which usually disappears even at much higher spiking levels. most compounds work at 2.5 ng on-column, though i have to fortify the mix to 5 or 10 ng for a few of the phenols, anilines, hexachlorocyclopentadiene, benzidines.
i tried spiking each of these 2 compounds by themselves at 10 times higher concentration. this resulted in ~60% recovery, and the c-gram seemed to show a trace of a possible reaction product (ms spectrum did not yield a good library match.)
thanks for any insight regarding this puzzle!