Advertisement

Communication error Agilient 1100 running chromeleon 6.8

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hi All,
I am currently using an Agilient 1100 running chromeleon 6.8.
I have just received this error message.

[Error] 17:28:12 {Pump} Communication error: LICOP command not acknowledged.
(pump on rejected while leak or shutdown or error )

I have been using the HPLC for over a year now and this is the first time I have seen this error message, do you have any ideas on how to resolve this ?
I currently cannot operate the HPLC.

Thanks,
Chemiger
99 problems and chromatography is every one !
You have to dry the leak sensor of the pump, then reset the pump (in Chromeleon menu "Control -> Command... -> Pump -> Reset, or turn off and on the pump)
2 posts Page 1 of 1

Who is online

In total there are 24 users online :: 2 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Baidu [Spider] and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry