Antibody analysis with size exclusion

Discussions about gel permeation chromatography / gel filtration chromatography / size exclusion chromatography

7 posts Page 1 of 1
I am working with an antibody. It seems like that the antibody is not stable in the mobile phase (diluent) during the size exclusion analysis (autosampler temp: 2-8C). Mobile phase is a 10 mM Sodium Phosphate, 700 mMSodium Chloride, pH 6.1 ± 0.1.
I am not able to run more than 12 injections of the antibody between system suitability injections. Any suggestions on how I can improve the stability of the antibody in the mobile phase so that i can run more samples in one sample set?

How does this instability manifest itself? Please describe what you’re seeing.

Best Regards
Learn Innovate and Share

Dancho Dikov
I am working with an antibody. It seems like that the antibody is not stable in the mobile phase (diluent) during the size exclusion analysis (autosampler temp: 2-8C). Mobile phase is a 10 mM Sodium Phosphate, 700 mMSodium Chloride, pH 6.1 ± 0.1.
I am not able to run more than 12 injections of the antibody between system suitability injections. Any suggestions on how I can improve the stability of the antibody in the mobile phase so that i can run more samples in one sample set?
There is a decline in aggregate level within a run(same day). It looks like sample degrades over a 15 hrs period.

It seams to me that you’re “problemâ€
Learn Innovate and Share

Dancho Dikov
Thanks danko. I will play with the pH and salt concentration.

Well, the question is: what is an aggregate? If the aggregation is reversible within the time frame of a day, it is not a permanent aggregate. The people in your formulation department might be interested in your findings. Maybe a change in formulation could prevent aggregation?

Is this an IgG? I have worked with several commercially available Mab of this type and have never seen a problem with aggregation. From this I assumed that handling these is well established for analysis and for the purpose of their use.
Could it be that your “aggregationâ€
7 posts Page 1 of 1

Who is online

In total there are 8 users online :: 0 registered, 0 hidden and 8 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 8 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry