				Halocarbons
				Posted: Thu Oct 31, 2019 6:42 pm
				by georgij23
				Hello!
Any body have experience with halocarbons (freons) testing. The question is it is possible check halocarbons purity  with GC FID detector? I find from agilent column guide parameters their is GC gas pro column and detector MSD.
			 
			
					
				Re: Halocarbons
				Posted: Sat Nov 02, 2019 12:08 pm
				by Pimi
				Hi!
If your baseline is really clen and stable it is possible to check purity of halocarbones but with low reliability.
For example. I check residual organic solvents in substance. One solvent is chloroform at 6 ppm. So I can see around 4 - 6 ppm of chloroform quantity, less amount only qualitative definition because signal/noise becom less then 10. 
So make model mix and check what the smallest amount you can see. 
PS: FID for halocarbons is not good desicion. ECD good )
			 
			
					
				Re: Halocarbons
				Posted: Sun Nov 03, 2019 8:05 pm
				by georgij23
				thanks, for your answer. I need check freons R410A, that is mix of difluormethane and pentafluormethane, and 50/50 percent mix. I dont need check small concentration ppm. That i ask about FID detector