SPE Separation with a Vacuum Manifold.
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
9 posts
Page 1 of 1
I am currently using a 12-port SPE vacuum manifold. This is very time consuming to manually adjust the flow rate. Does anyone know of a semi-automatic or automatic system that will produce a constant flow rate?
-
- Posts: 2916
- Joined: Mon Aug 30, 2004 10:19 pm
I assume that you adjust the flow with a vacuum gauge and a throttling valve. Once it is set, it should work for subsequent applications, right?
-
- Posts: 77
- Joined: Tue Apr 22, 2008 2:11 pm
Uwe,
From my experience of SPE this isn't the case. The density of the packing varies from cartridge to cartridge which has an effect on the flow rate as does the amount of particulate being drawn from the samples.
John,
I don't know of any automatic methods for this. I agree that it is a very time consuming process.
Rich
From my experience of SPE this isn't the case. The density of the packing varies from cartridge to cartridge which has an effect on the flow rate as does the amount of particulate being drawn from the samples.
John,
I don't know of any automatic methods for this. I agree that it is a very time consuming process.
Rich
-
- Posts: 2846
- Joined: Mon Aug 30, 2004 7:17 am
We put a little faucet (Luer lock connections) on each cartridge to regulate them individually. There are metal as well as plastic versions available, I don´t remember from where we got them, but I seem to remeber that the plastic versions have a tendency to bind (get stuck). SPE manifolds were used extensively for RIA and ELISA.
-
- Posts: 47
- Joined: Wed Feb 09, 2005 7:48 pm
It's two well known (and working well) "automatic" SPE systems :
1. Gilson - GX 271 ASPEC http://www.gilson.com/Products/product.asp?pID=281
GX 274 ASPEC http://www.gilson.com/Products/product.asp?pID=282
2. Caliper
Rapid Trace SPE Workstation http://www.caliperls.com/products/rapid ... tation.htm
We use old Gilson system. Constant flow rate is obtain by air pressure from the needle. You can easy create you own methods with different parameters like flow rate, time of drying, time of loading of the sample etc. by software controlled the system.
Both system are quite expensive and (in my opinion) it is reasonable to use it only with high number of samples.
1. Gilson - GX 271 ASPEC http://www.gilson.com/Products/product.asp?pID=281
GX 274 ASPEC http://www.gilson.com/Products/product.asp?pID=282
2. Caliper
Rapid Trace SPE Workstation http://www.caliperls.com/products/rapid ... tation.htm
We use old Gilson system. Constant flow rate is obtain by air pressure from the needle. You can easy create you own methods with different parameters like flow rate, time of drying, time of loading of the sample etc. by software controlled the system.
Both system are quite expensive and (in my opinion) it is reasonable to use it only with high number of samples.
-
- Posts: 366
- Joined: Wed Feb 02, 2005 9:34 pm
Some vendors sell SPE cartridges where you can push the liquid through with a syringe, instead of using the approach of pressurizing to push the liquid.
I have to admit that I have never tried this, but it sounds like a good option if you are doing something like 7 or 8 samples at a time. For situations where you have 40 samples at a shot (or more) the pressurized manifold is probably the better way to go.
I have to admit that I have never tried this, but it sounds like a good option if you are doing something like 7 or 8 samples at a time. For situations where you have 40 samples at a shot (or more) the pressurized manifold is probably the better way to go.
-
- Posts: 366
- Joined: Wed Feb 02, 2005 9:34 pm
To clarify: push through with a syringe piston.
-
- Posts: 17
- Joined: Sat Sep 13, 2008 11:29 am
You must have individual valves for each column and manually close off those columns that drain slightly quicker. I don't know how you could possibly have something "automatic" to shut off a particular column until it runs dry and then it's too late. I use a 24 port all the time but usually limit the number of samples to one side of twelve and pay attention.I am currently using a 12-port SPE vacuum manifold. This is very time consuming to manually adjust the flow rate. Does anyone know of a semi-automatic or automatic system that will produce a constant flow rate?
-
- Posts: 22
- Joined: Wed Feb 09, 2005 10:29 pm
I agree wholeheartedly with Mr. Carver. See these links:
http://www.biotage.com/DynPage.aspx?id=24411#STOPCOCKS
http://www.varianinc.com/cgi-bin/prodin ... KMLIHHOJFL
Best regards,
Tom
http://www.biotage.com/DynPage.aspx?id=24411#STOPCOCKS
http://www.varianinc.com/cgi-bin/prodin ... KMLIHHOJFL
Best regards,
Tom
Thomas J. Waeghe
MAC-MOD Analytical, Inc.
Chadds Ford, PA 19317
800-441-7508
twaeghe@mac-mod.com
www.mac-mod.com
MAC-MOD Analytical, Inc.
Chadds Ford, PA 19317
800-441-7508
twaeghe@mac-mod.com
www.mac-mod.com
9 posts
Page 1 of 1
Who is online
In total there are 11 users online :: 1 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 10 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 10 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science