There is an outside possibility that my injection technique (or the low injection volume?) is causing the problem but I don't think so.
I'm doing a analysis of the amount of an organophosphate absorbed by a porous solid in both water and 95% Ethanol/H2O. No internal standard. My questions are is my method improvable and what RSD would it be realistic to achieve?
The instrument, column, and method info:
Shimadzu GC-17A chromatograph w/ JW Scientific DB-5 column (30m, .25 mm, .25 um). I am using a plunger-in-barrel 10uL syringe with standard (NOT on-column) manual injection.
Injector: 250 C; 100 kPa; total flow rate 104 mL/min; column flow rate 1 mL/min; linear velocity 26.9 cm/sec; split ratio of 100:1 w/ a 1 uL injection volume.
Temp. Program: 100 C for 1 min, then increase by 20 C/min for 5 min, then hold at 200 C for .5 min. My analyte elutes out just after 5 minutes.
I know it could be blowback or uneven vaporization in the liner. But I also figured that my pressure or temp or type of injection may be responsible. Wouldn't ethanol analysis benefit from a lower injector temp or higher pressure, etc. than water?
Thanks so much for your help and sorry for the huge post!!
P.S. On another note, I am having a hard time getting tiny air bubbles off of the white cap (gas-seal?) on the tip of the plunger while working with the water solution. I invert the syringe and flick it over and over but the bubble won't come free. How can I fix this?