SO2 question
Posted: Thu Feb 05, 2009 7:04 pm
Can anybody recommend a capillary GC column for SO2 test ? (our GC can only connect to capillary column).
We need to test SO2 concentration in air stream in ppb level up to ppm level, using GC with FPD detector. The stream has moisture and is relative pure.
I tried regular general purpose capillary column but the results bounce around with +-15% error for a standard SO2. I also found that if I quickly introduce SO2 samples of fixed concentration into GC one by another, the following peaks will show higher intensity and areas.
I'm trying to fingure out what cause this error. It can only be column or tubing, as I'm using a Sulfnert sampling loop (100 ul) from Restek.
Thanks
We need to test SO2 concentration in air stream in ppb level up to ppm level, using GC with FPD detector. The stream has moisture and is relative pure.
I tried regular general purpose capillary column but the results bounce around with +-15% error for a standard SO2. I also found that if I quickly introduce SO2 samples of fixed concentration into GC one by another, the following peaks will show higher intensity and areas.
I'm trying to fingure out what cause this error. It can only be column or tubing, as I'm using a Sulfnert sampling loop (100 ul) from Restek.
Thanks