I am currently trying to assay Dextran 70, a ~70kD sugar molecule by SEC with detection by RI. The standard has a small peak at ~5 min and one main peak eluting at ~7.5 min when dissolved in water. However, when analyzing a sodium chloride injection solution of Dextran 70 at the same concentration, there is a large fronting peak under Dextran 70 (approx. 10% of the STD area). The placebo is 0.15M NaCl and this peak is not present in water injections. Fresh preparation of the NaCl solution has eliminated contamination issues.
Here is a review of my system:
Thermo Separation Systems Pump/Autosampler
Waters 2414 RI (34C internal temp, 128 attenuation)
Column - Phenomenex PolySep-GFC-P4000 300mmx7.8mm
MP - Water
Flow Rate - 1.0mL/min
Injection - 20uL (200uL loop)

So does it sound like we have the wrong column for this analysis and neither of these are really being retained? Doing a rough calculation, the column should have approx. 7mL and that would equate to 7 min of mobile phase - close to my RT for both. Or am I missing something else? If it is the wrong column, what columns would work for this? Or could a guard column simply remove the sodium from the equation? Any help would be much appreciated, thanks!