wavelength accuracy

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Hi to all,

I would like to know
what concentration of standard solution should be used to perform the wavelength accuracy test with aqueous caffeine solution for HPLC-UV-Vis detector calibration and is there any guideline to use caffeine as standard.

Thanks in advance

Nageshwar

I would favour the use of a STD which does not produce an absorbance in excess of 1000 mAU (1 AU) at the theoretical caffeine absorbance maximum.

We use a 25 µL injection of a 50 µg/mL STD solution.
Good judgment comes from bad experience, and a lot of that comes from bad judgment.

you should check if your detector has a mercury lamp.
this lamp has a set of very specific WL and it is used these days in many detector for WL accuracy of the optics

Hi,

The detector is built with duterium lamp.

Thanks

Nageshwar
4 posts Page 1 of 1

Who is online

In total there are 26 users online :: 2 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot], Semrush [Bot] and 24 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry