HPLC retention modelling

Posted: Wed Oct 23, 2019 2:40 pm
by MaxSpecs
Trying to understand mechanics of HPLC modelling (like used in DryLab and other software) ...
In fact I even tried to code a simple routine in Matlab to do that.
I'm not clear though on one aspect of the numerical approach to finding the solution based on two training gradient runs (I'm talking about a straight one parameter modelling - %B change).
Assuming I have a model equation lnk'=f(B) with say two parameters (a,b) is the best approach to just numerically simulate 2 gradient runs (identical to the training runs) as a sequence of small isocratic steps using different a and b for each pass (iteration/optimization) until convergence is found between simulated and real tR on both gradients?
It kind of works but seems computationally expensive...
Is this how the commercial softwares do it?

I'm a bit confused by reports of equations linking lnk to the elution time in the simulated gradient as reported by some software...

Re: HPLC retention modelling

Posted: Fri Oct 25, 2019 2:09 am
by tom jupille
Essentially, the commercial programs like DryLab use retention to estimate k* (the average k'). Given two gradients with different slopes, that is used to estimate retention for other (usually intermediate) slopes. The actual algorithms used are laid out (in gory mathematical detail) in the Snyder and Dolan book on gradient elution. Here's the link on Amazon:
https://www.amazon.com/High-Performance ... 471706469/