Hi

Does anyone know of sources in Australia for ACN. Crown Scientific our usual supplier have just qouted me 6 months when I rang up to find out how far away my order I put in 6 weeks ago.

I now have 1 L left in the lab. And we have just started up production on one product again that I need it for sample prep.

If not any reliable (consistant quality) sources in China, we import all our raw materials directly from there anyway

Regards

Kerrie

I just wanted to add my 2 cents to using solvents other than ACN. We have some OTC pharmaceutical products where our ACN-containing mobile phase had the API elute between fragrance ingredient #1 and fragrance ingredient #2.

When we removed ACN and substituted a similar percentage of methanol, API eluted AFTER fragrance ingredient #2 and before fragrance ingredient #3. Yes, the pressure was considerably higher as well, and we did raise column temperature a small amount.

When we next removed methanol and substituted a different solvent, API eluted AFTER fragrance ingredient #3 major component and before fragrance ingredient #3 minor component. We will begin re-validating soon using one of those options. Sample preparations do not use ACN and will remain exactly the same.

For a non-regulated product we're trying to convince our QA department that we should only need to demonstrate equivalency of the chromatography, since the sample preparation will be the same, but they still want a zillion sample preparations done, natch, not just showing the chromatography delivers equivalent results.....love that QA.......
Equivalent percentages of methanol vs. acetonitrile or another solvent don't necessarily mean equivalent eluent strength. In order to achieve this, you would need to consult a nomogram like this (sorry for the small size...it's the only one I could find on short notice).

Image

The top is %ACN, middle is %MeOH and bottom %THF. Selectivity might not be the same, but this will get you closer than simply switching 50% MeOH for 50% ACN or whatever the composition...according to the chart, 50% ACN would be 60% MeOH or so, for example...good luck!

CJ Rassbach

One has to be very careful about these conversion rules. The reason is that the solvent selectivity affects the conversion significantly, and so does the stationary phase.

Here are some published equations (with the authors), unfortunately in a form that is explicite for acetonitrile. The equations describe the relationship between the volume fractions of the different solvents:

Haddad:

MeCN = 0.9*MeOH^2 - 0.21*MeOH + 0.126

Schoenmakers:

MeCN = 0.23*MeOH^2 + 0.57*MeOH

In my research I found that the equation depends also on the type of bonded phase

C18 and C8:

MeCN = 0.25*MeOH^2 + 0.63*MeOH

EPG phases:

MeCN = 0.8*MeOH

Please remember that all these equations are just rules of thumb, which will vary with your analytes, especially if you have ionic analytes.

I'm not having much luck eliminating acetonitrile from my prep and analytical methods either. Unfortunately, it works quite well compared to other solvent systems!

Currently we are analyzing and purifying a variety of peptide sequences in 0.1% TFA mobile phase. The best I seem to be able to do is 2:1 MeCN:IPA in the B phase. Any greater, and the backpressure becomes unacceptable. Methanol doesn't seem to be an option in our case, as I am seeing methyl ester formation in peptides and building blocks unless I go with 0.01% TFA or 0.1% formic acid. Now we are hearing the shortage could persist through 2009... fantastic!

I put the arm on John Dolan to record a "mini-seminar" that summarizes some of the issues. You can view/listen to it here:

http://www.lcresources.com/more_resourc ... p?f=3&t=20

Note that you will have to register (free) in order to access it.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
Hello,
We are supplier of HPLC acetonitrile from China.Our spec is below:

ASSAY(BY GC) % MIN 99.9
WATER (BY KF) mg/kg MAX 100
RESIDUE AFTER EVAPORATION mg/kg MAX 2.00
TITRATABLE ACID meq/g MAX 0.0008
TITRATABLE BASE meq/g MAX 0.0006
UV ABSORBANCE (1.00-CM CELL VS WATER)
AT 190 nm MAX 1.00
AT 200 nm MAX 0.05
AT 210 nm MAX 0.03
AT 220 nm MAX 0.01
AT 254 nm MAX 0.005

I don't know why the shortage situation is so bad in the world market.If you interest in our product.Please contact me at jorry@hnharvest.com
Hi

Does anyone know of sources in Australia for ACN. Crown Scientific our usual supplier have just qouted me 6 months when I rang up to find out how far away my order I put in 6 weeks ago.

I now have 1 L left in the lab. And we have just started up production on one product again that I need it for sample prep.

If not any reliable (consistant quality) sources in China, we import all our raw materials directly from there anyway

Regards

Kerrie

As far as I know, after talking to two people who buy ACN from China and Russia, the quality is not the same as required for HPLC grade. They are doing THREE distillation two purify ACN from these two countries to get it to the desired quality......plus the price of the raw ACN is not that great.
I think that situation is getting better, two month ago I had offer of $850 for case of 4x4L and two weeks ago it is already $650, still way above $165/case we used to pay.......but we managed to buy ACN from the lab which was going out of business for only $120/case...got lucky and now have ACN for another 7-9 months!

P.S. Plus none of the pharma companies are going to buy 12 MT from unknown supplier of the grade you offer. It is better to approach either reselers or manufactureres with distillation capabilities to bring your ACN to desired spec.
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com

I know of a compnay that paid lots for a 250 litre drum of 99.9% ACN only to find it was 95%, I would stay well clear.
:roll: :roll: :roll: :roll: :roll: :roll: :lol: :lol: :lol: :lol: :lol:

Hi,

I never had much faith in those conversion charts to start with so 50% ACN (no TFA) which I was using converts to 40% MeOH . REALLY.

For tebuthiuron (herbicide) this 40% MeOH gives unusable really really broad peaks. 50% MeOH gives broad peaks (which I wouldn't use) with elution at 12min not the 3.0 min for 50% ACN.

At 60% ACN plus 0.1%TFA to sharpen the peaks up I get nice clean peaks at 5.7 min. I might use this or maybe play a little more to reduce the retention time.

So use them with caution.

:cry: :cry: Still no ACN. So I look like I might be doing a bit of validation work.

Kerrie
Why did you select IPA over Methanol?

Also, has anyone tried going to 3.0 mm ID columns? If so, was it very difficult?

Thanks
Hi Paladin - the data (TI475E) I posted earlier gives some information on this.
Assuming column length is kept the same:

Cross sectional area of cylander = pi(r^2) = pi[(0.5d)^2]
For 4.6mm: Area = pi[0.5x0.46cm)^2] ~ 0.166 cm^2
For 3mm: Area = pi[0.5x0.3cm)^2] ~ 0.071 cm^2

Ratio of c.s. area for 3mm/4.6mm = 0.071/0.166 ~ 0.43

So, if flow rate is 1mL/min (on 4.6mm I.D.), new flow rate should
be reduced proportionally to 0.43 mL/min.

Extra column volume effects may need to be considered.
And you may need to reduce inj. volume by 1/2
Could you give any recommendation to what extent a re-validation is required for a HPLC method, when you change the ID of a column, and consequently flow as well as injection volume?

These conversion charts have to be based on some type of experience. From my equations above, you will always end up with more methanol than acetonitrile, except for the Haddad conversion in very high water content, for which it may not apply anyway. From my equation above, 50% acetonitrile will require between 60 and 65% methanol for equal retention.
Why did you select IPA over Methanol?

Also, has anyone tried going to 3.0 mm ID columns? If so, was it very difficult?

Thanks
Hi Paladin - the data (TI475E) I posted earlier gives some information on this.
Assuming column length is kept the same:

Cross sectional area of cylander = pi(r^2) = pi[(0.5d)^2]
For 4.6mm: Area = pi[0.5x0.46cm)^2] ~ 0.166 cm^2
For 3mm: Area = pi[0.5x0.3cm)^2] ~ 0.071 cm^2

Ratio of c.s. area for 3mm/4.6mm = 0.071/0.166 ~ 0.43

So, if flow rate is 1mL/min (on 4.6mm I.D.), new flow rate should
be reduced proportionally to 0.43 mL/min.

Extra column volume effects may need to be considered.
And you may need to reduce inj. volume by 1/2
Could you give any recommendation to what extent a re-validation is required for a HPLC method, when you change the ID of a column, and consequently flow as well as injection volume?
Hi Sahav -

Sorry, I'm the wrong person to ask (I no longer work in a regulated environment).
But even in the (proposed?) General Chapter <621> Chromatography,
I think the recommended limit was to be +/- 25% for column I.D.
So - I imagine changing from 4.6 to 3mm I.D. would require re-validation.

This is just a guess though.
But even in the (proposed?) General Chapter <621> Chromatography,
I think the recommended limit was to be +/- 25% for column I.D.
So - I imagine changing from 4.6 to 3mm I.D. would require re-validation.
This is just a guess though.
Columns are generally avaiable in inner diameters of 4.6mm, 3.0mm, 2.1mm and 1mm, if I'm remembering correctly. So following the guideline of +/- 25% for column I.D., no other i.d. would be acceptable. So that recommended limit is pretty worthless/useless/ridiculous, didn't anyone who actually performed HPLC have a hand in it?

Interesting, for GC, the i.d. limits are greater, so those actually can cover a column that's commercially available.

To keep the topic alive:

FDA point of view:
http://www.fda.gov/cder/ops/Acetonitril ... pdated.pdf

Ace
I am sitting on 6X4 liters of HPLC grade (J.T. Baker), unopened, that I have available at cost (our lab budget was trimmed recently). I am in the St. Louis area. I had over-ordered a couple of years ago for a method that was never adopted. End user must be a qualified lab with a legitimate purpose and preferably a real need. I will not ship, but will consider delivery within about a 100 mile radius.
125 posts

Who is online

In total there are 27 users online :: 1 registered, 0 hidden and 26 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 26 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry