GOM » Fri Jul 08, 2016 10:10 pm
Hi Joe
I posted the tubing today.
I prefer to use it in thermal desorption mode.
The tumbling SPE or diatomaceous earth suggestions sound better for solvent extraction.
I recall conditioning the tubing in an empty 1/4" glass GC column - several of the pieces placed in and left with a small flow of helium at high temp to remove volatiles and reduce siloxanes. You could use your thermal desorber for this stage.
Below - The top thermal desorption trace is one that I ran to compare a poorly conditioned tube with that of a better conditioned tube (lower trace). The major peaks are siloxanes
If you must try solvent extraction try 75/25 methanol/chloroform. CHCl3 on its own seems to just get totally absorbed and swell the tubing. Try testing a few pieces first to get the right solvent mix.
Sometimes it is helpful to think of PDMS as being a very thick low polarity liquid/solvent.
It would be worth preconditioning the tubes with a triple extraction using that solvent. Say 200uL for each extraction using one of those vials that you used to show us a picture of the extracted mites. Try running each of those 3 conditioning extracts in reverse order to see if the background siloxane peaks are reduced to an acceptable level
This is a comparison trace of a sample (top) vs blank (bottom) of a solvent extracted tube that I used in the analysis by exposure to fungal volatiles
In the case of a negative result on your plates
This would be interesting in itself.
In my opinion, every negative result is actually positive in that it directs you towards the correct result.
You said originally that an extract had exhibited biological activity.
1. Even if the TLC solvent system had not migrated and separated your applied spot then one would have expected some activity around that spot .
2. or even you had just applied it and not taken it through the procedure. as a control
Unfortunately I have no experience of how you confirmed the biological activity of your extract.
The implication from 1 and 2 could be that the active compound(s) are too volatile and are being lost in the TLC procedure
Regards
Ralph