THERMO, TRACE GC ULTRA WITH PROBLEM
3 posts Page 1 of 1
I have a Trace GC ultra coupled to a TSQ-XLS, from Thermo.
My Trace GC ultra is not working because have an electronic board burned. The electronic board is the ANALOF FEEDBACK PCB, Part number 23648340. I am trying to find anyone that have a old board.
I think that the problem of my GC is related with that, because when I turn ON the instrument the screen show me the next message MB EEPROM IS BAD.
Can anyone give me a suggestion please?
No country info. Did you try Thermo ? Are you sure that swapping the board will not result in the same board burning out again as there could be an issue on the power supply ? CE Instruments can service these GCs and we have some stock boards to support our customers. I guess however that you are not UK based ?
More info. required :wink:
Is this a PTV injector system?
If so solvent valve can be the cause of burning the board
3 posts Page 1 of 1
In total there are 27 users online :: 0 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Users browsing this forum: No registered users and 27 guests