unmgvar » Sun Nov 13, 2005 7:51 pm
Uwe,
>2 particle size columns can't be SEC colunms, they don't have the geometry for that. monolith columns on the other hand have it even thou it is not their main "separtion power". >2 particle size do the work throu absorption only. i do not mean that they cannot separate proteins and peptides (they do it with absorption) but, in comparison monolith column do it better i believe.
If detector's rate are not of an issue then why do Agilent and Waters introducing a new 80Hz detector? the detection issue is critical for both small particle size and monolith columns. as you decrease peak width, increase flow rates or both at the same time, you need to increase detection power. hence the need for faster detectors. i wonder thou the S/N they get now. Saw a similarity check done by agilent in one of their advertisement. i would never have aproved such a check to one of my analysts. they must have used all the "tricks" in the book to make this one stick.
Anyway, as Tom Jupille reminded us, the "battle" is basically between monolith columns to >2u particle size columns. smaller particle size is just more of the same. Monolith are far more innovative (doesn't make them better of course).
in the long run it is should be by far easier to polymerise a greater surface area in the same column i.d. then to have to press it.
monolith columns also provide us with a greater internal dwell volume and therefore less back pressure. exactly the problems that >2 particle size column have.
and with all that said i will not get started on HPTLC. which is also a very good solution for all of our troubles. especially in saving time and solvent consuption, and cost per batch. we are just to much in love with our HPLC's to even look that way.
Newman768 ,
you must know of course that at those flow rates according to the Van-Deemter plot your >2 particle size column performs barely better then a 3u column. you need higher flow rates in order to get better results. i am guessing that your column is too short and that your product gets out too soon. but then the longer the column the higher the pressure.
Alex Buske,
you wrote: Having 50000 or so plates would be worth it. Even at higher costs.
then why don't all QC labs have a few MS in them?
Costs per batch are very much a factor for opting a certain technology. >2 particle size instrument increase the costs in general even if you can show that in some points they do save money. the overall picture is costlier.
Monolith column also have a big drawback with the huge increase in solvent consuption.
and still i hold my bet. in the near future, people will be connecting monolith column to UPLC instruemnts.