GC 5890 FID/Purge and Trap Baseline Drift Problem

Discussions about GC and other "gas phase" separation techniques.

47 posts Page 4 of 4
You have a 1/16 inch swagelok at the end of the transfer line. The only extra materials you need to connect the column to the transfer line is a 1/16 inch swagelok nut and a 0.8 mm ferrule. You work with 0.53 mm i.d. columns, so you must have 0.8 mm ferrules somewhere, and I would be surprised if there is not a 1/16 inch swagelok connector somewhere as well.

The most likely cause for the contamination is the inlet septum. By making a direct connection between transfer line and column you eliminate not only the septum but also the dead volumes in your current set up.

Despite what your supervisor says, you should have a purifier in your carrier gas line.

Peter

You were very right Peter. I have run my GC with my inlet disconnected. I got a flat baseline. This is my chromatograph
Image
I am running a RT Marker right now to see what the effect of having no inlet will be.
I am not seeing any peaks at all with my RT Marker. But I also do not have any column head pressure reading without my inlet connected. Not sure why?
47 posts Page 4 of 4

Who is online

In total there are 47 users online :: 0 registered, 0 hidden and 47 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 47 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry