HPLC Method for sodium acetate anydrous determination

Discussions about GC and other "gas phase" separation techniques.

63 posts Page 4 of 5
You have headspace ? If yes - not need liquid/liquid extraction.
In headspase you have liquid/gase phase extraction , very effective by acetic acid water solution. Acetic acid is high boiled, but vary volature liquide.
The sulfuric acid (in water solutions) not volature liquid. Quicly dead of column by cross-link phase will not. Effectivity of column , tailing factor - in you quest is secondary question.

I'm recomendate. Make the some water solution of acetic acid by concentration 10, 5. 1, 0.5, 0.1 mg /ml and analise it (headspace 80-90 C, 10 ml , after add 1-2 drop of concentrated sulfuric acid and Na2SO4 ). For (from) determerminate calibration curve linearity and sensivity of analitical method.
After that we will discuss results.
You've say "The sulfuric acid (in water solutions) not volature liquid". Can you explain to me why you've say it? I want understand. Thank you very much!

100,0% of acid remain in Head space vial? Some part vapour don't vaporize?

Thanks!
You seem to have had plenty of good advice for this post. But if you still need some advice please let me know. We are Chromatography specialists, willing to give technical support if required. We supply chromatography consumables, including Columns, 30 years experience of chromatography may just give you some assistance. Don't apologise for your English, it's we English people who should apologise for not making enough effort communicate in other languages. Have a look at our website http://www.greyhoundchrom.com
Good luck with your research
Thanks for your interesting and for your understanding! I'll see your website!

Have a Good day!
If you can see "parts of compounds" in the hexane layer then they are not in solution, they are particles in solution. Try filtering the hexane layer through a syringe=tip filter after you do the extraction and before you inject the hexane layer onto the GC. Despite what Carl says you can cause all sorts of problems by injecting dirty samples.
Peter
Trying to keep things simple for this new user during these range finding experiments. Every sample is dirty, some worse than others. My undertsanding of the concern was sample dissolved in the hexane, not particulates. The user will have to clarify this.

Also, I'm still not convinced the LLE approach will work since acetic acid has higher solubility in water than hexane.

I'm glad to see DSP reviving the headspace discussion.

If using headspace I recommend using a single solvent that completely dissolves the sample such as DMF, DMSO, etc, to avoid baises caused by incomplete extaction of acetic acid from the core of particles. If using sulfuric acid (BP ~330 C) a small amount of acid will be in the vapor phase but this should not cause major problems with the GC column.
A. Carl Sanchez
Hi Carl

If alemaggot can see compound in either the water or the hexane then what he is seeing cannot be in solution (unless his compound is coloured of course), it has to be particles in suspension. Why inject a dirty sample when there is an easy way to clean it up ?

Discussion of headspace methods is only useful if alemaggot has a headpsacer.

Peter
Peter Apps
Yes guys, I've a head space sampler ;)

Than... You advise me to use this method? Ok. You're all agree with kind DSP007 says?
Headspace would be great. Weight your sample into a vial, add sulfuric acid and run.
I suggest using a solvent mixture that completely dissolves your solid sample such as 90/10 DMF/1M aq H2SO4 (v/v) or 90/10 DMSO/1M aq H2SO4. Heat at 80C for 15min prior to injecting the headspace.
A. Carl Sanchez
You have headspace ? If yes - not need liquid/liquid extraction.
In headspase you have liquid/gase phase extraction , very effective by acetic acid water solution. Acetic acid is high boiled, but vary volature liquide.
The sulfuric acid (in water solutions) not volature liquid. Quicly dead of column by cross-link phase will not. Effectivity of column , tailing factor - in you quest is secondary question.

I'm recomendate. Make the some water solution of acetic acid by concentration 10, 5. 1, 0.5, 0.1 mg /ml and analise it (headspace 80-90 C, 10 ml , after add 1-2 drop of concentrated sulfuric acid and Na2SO4 ). For (from) determerminate calibration curve linearity and sensivity of analitical method.
After that we will discuss results.
You've say "The sulfuric acid (in water solutions) not volature liquid". Can you explain to me why you've say it? I want understand. Thank you very much!

100,0% of acid remain in Head space vial? Some part vapour don't vaporize?

Thanks!
OK
Smell the first aqueous solution of sulfuric acid, then an aqueous solution of acetic acid.
In the first case there is no odor - sulfuric acid in an aqueous solution of non-volatile. In the second case you smell even a 1% acetic acid, acetic acid is volatile.

Likewise, if HS will enter into GC-acetic acid will be present in the vapor phase. A sulfuric acid is in liquide phase.But sulfuric acid remains in the liquid phase and does not get in the instrument .
Before you try sodium acetate you need to test whether or not you can see acetic acid when that is the only component in the sample.

Make up a 1% solution of acetic acid, and adjust you headspacer and GC settings until you can see a nice sharp peak that is not present when you analyse clean water. The halve the concentration of acetic acid and see whether the peak gets smaller. If it does it is probably acetic acid.

From the amounts of sodium acetate that you used it seems that the concentrations that you are expecting in your samples might be rather low. Can you confirm what concentrations you are expecting ?

Peter
Peter Apps
Hi guys!

Tomorrow I've try the injection.

I've prepare two vials with about 1 g of sodium sulfate, 5 ml of sodium acetate solution 0,01 mg/ml in water. In one of this vials I've add 2 drop of sulfuric acid.

In the result of vial n 1 without sulfruic acid I've found about 41000 area units. In the second results, with acid, I've found about 27000 area units. Then is less.

How can we explain it?
Before you try sodium acetate you need to test whether or not you can see acetic acid when that is the only component in the sample.

Make up a 1% solution of acetic acid, and adjust you headspacer and GC settings until you can see a nice sharp peak that is not present when you analyse clean water. The halve the concentration of acetic acid and see whether the peak gets smaller. If it does it is probably acetic acid.

From the amounts of sodium acetate that you used it seems that the concentrations that you are expecting in your samples might be rather low. Can you confirm what concentrations you are expecting ?

Peter
Hi Peter!

In my product the sodium acetate limit is less than 0,02%. Usually there is a small residue of it after production (about 0,003 %)
[quote="alemaggot
Hi Peter!

In my product the sodium acetate limit is less than 0,02%. Usually there is a small residue of it after production (about 0,003 %)[/quote]

Whoops ! that is rather low. Acetic acid is difficult to do by GC at low levels. Maybe we should have found out the target concentrations four pages ago (or did I miss it before ?).

Let's see what the experts think.

Peter
Peter Apps
Guys have you some news?
Guys have you some news?
mark
Guys have you some news?
mark
what?
63 posts Page 4 of 5

Who is online

In total there are 20 users online :: 2 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry