Thanks Ron. That's the direction I was leaning. I think you just pushed me over the line. The discrete dynode EM costs about $200 more than the cheapest channeltron that I can find (and about $400 less than the upper end of the range). All it would take would be a few extra months of life to make up the difference.

Hi Don, you may want to take a look at http://csa-service.com/store/ they are contracted by Agilent to provide service/qualifications on 5890s and Agilent sold them everything they had as far as official 5890 parts.

DCM, deionized water/HPLC water, hexane, and methanol are great for source cleaning. I would do your last rinse in methanol as it is miscible with any remaining water and has minimal evaporative residue.
47 posts Page 4 of 4

Who is online

In total there are 28 users online :: 1 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: dlntx9 and 27 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry