i run USP standards and then the samples.
The mobile phase is made in bulk by individually measuring water(69%), methanol(28%) and glacial acetic acid (3%). The pH of this is somewhere around 2.4pH
Problem starts with that when i start the HPLC machine everyday, with the same mobile phase, and equilibrate it for 2 hours, my pressure stabilises at different pressures everyday. But, once it stabilises, it remains the same throughout the day-- so there is an interday variability in the retention times with change of pressures.
Also, i have noticed that when i run a set of 45 injection, the peak that comes at 6 min, after 45 inj, it comes at 4.5 min near about. And the pressure is stable!
The solutions are filtered before injection. The column is washed after every 40-50 injections with 100% methanol. And i have had 4 new columns of different kinds just to see if columns are the source are problem.
I have checked the HPLC machine, it is OK.
Cant think of any thing else..