ym3142
each day, with a fresh system, peak area is 19-20, and then it decreases with multiple injections of standards till at the end of the day, it is 14. So I believe normal should be 19.
With a fresh system, you mean fresh mobile phase?
If you make a bottle mobile phase (A), inject a few standards from the same vial (A), your area decreases. What if you use the same standard solution (same volumetric flask), but in another vial (B)? And a freshly prepared standard in vial (C)
And if you make mobile phase fresh (B) and inject again vial (A)?
I would suggest doing the following:
Make mobile phase A, make a standard in vial A, inject a few times, within a couple of hours (4-5) this should be enough to see the decrease I think. Then fill a new vial (B) and inject. Also prepare a new standard solution and inject this vial (C).
Now make fresh mobile phase (B) and inject vial A, B and C.
Do the whole test within 1 day, and you can see if it's vial related, stability of the solution, mobile phase related, or something else.
Just my suggestion.....
Bart