I'd agree with UN. I compared a few of those for polar compunds a while beck and had the most luck with Phenomenex Polar-RP (Phenyl column with embedded groups that tolerate 100% water).

As for the ZIC HILIC colum you mention I definitly think there are good possibilities to achieve good separation and retention on that column.
Svante Vikingsson
PHD student
Clinical pharmacology
Link

Danko,
irrespective of Sadilek´s coming results: I didn´t specify attraction, nor repulsion, but any dipole has the opposite formal charge to make it a dipole. The overall result seems to be an attraction, see as an example: RS Drago, Physical Methods in inorganic Chemistry, Reinhold, 1965, page 66.
But this is a mute argument here as these interactions seem to be washed out (overwhelmed) by mobile phase interactions.

Hans, forget the dipole-ion interaction. Think hydrogen bonding between protonated silanol (hydrogen donor) and the double bonded oxygen (acceptor). OK?

Best Regards
Learn Innovate and Share

Dancho Dikov

So, here is the result:

http://img187.imagevenue.com/img.php?im ... _158lo.JPG

As you can see, there is no difference between buffered phase and phase without buffer. I did two injections of sample for each phase, chromatograms are identical. I tested 2 columns (originally used Supelco and Symmetry Shield C18) without significant change.

Hi Sadilek,

Thanks for data. Shame it didn’t work, but I still think it was worth trying – for the sake of the knowledge.
Additional question: How did the chromatography look like with buffer at pH 3 – 4 and higher? Also, was the buffer concentration the same – both at low pH (i.e. 2.3) and at higher pH values?

Something else, not related, is intriguing me though: With the described column (250x4.6mm) and 1 mL/min flow rate, it shouldn’t be possible to see peaks eluting before 2.7 min (approx.) – dead volume. Nevertheless, there is a distinct peak eluting at 1.2 min, most distinct on the chromatogram bellow (just water/MeOH). Do you start collecting data a minute or so after injection?

Best Regards

And thanks for sharing with us.
Learn Innovate and Share

Dancho Dikov

Hi danko,
... How did the chromatography look like with buffer at pH 3 – 4 and higher? Also, was the buffer concentration the same – both at low pH (i.e. 2.3) and at higher pH values?
If I can remember, RT of principal peak was the same, buffer concentration approx. 15-20mM. Now I ordered recommended Atlantis T3 and Primesep B2 columns for additional testing.

... With the described column (250x4.6mm) and 1 mL/min flow rate, it shouldn’t be possible to see peaks eluting before 2.7 min (approx.) – dead volume. ... Do you start collecting data a minute or so after injection?
No, data collection is started immediatelly after start. Maybe used analytical column is well packed. As I can say from my experience, dead volume of 250x4.6;5um columns is usually about 1.8 - 2 mins.

Regards

For neutral compounds - you can try using H2O:THF (or a combination
of MeOH:THF). Get rid of the buffer.

Another column to try for neutral compounds is Cadenza CL-C18.
It excels at shape selectivity.

Below are steroid applications. Changes in selectivity were obtained by changing the solvent type:
http://www.silvertonesciences.com/files/TI324E.pdf

OK, this agrees with what we have seen: No effect. The hydrogen bonding is als swamped, apparently.

The hydrogen bonding is als swamped, apparently
Hi Hans,

“Swampedâ€
Learn Innovate and Share

Dancho Dikov

Danko, I didn´t say that there is no hydrogen bonding, I said that it is obliterated, if there is any, by mobile phase interactions. I am talking about equilibrium shifts toward the mobile phase. In analogy you can eliminate retention due to hydrophobic interaction by increasing the organic content of the mobile phase.... so "swamping" doesn´t eliminate hydrophobic interaction (or hydrogen bonding, etc.), it just shifts it around.

Hans, OK :)

Best Regards
Learn Innovate and Share

Dancho Dikov

Thank you very much for stimulative hints.

Regards
42 posts Page 3 of 3

Who is online

In total there are 16 users online :: 0 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry