As for the ZIC HILIC colum you mention I definitly think there are good possibilities to achieve good separation and retention on that column.
PHD student
Clinical pharmacology
Link
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
If I can remember, RT of principal peak was the same, buffer concentration approx. 15-20mM. Now I ordered recommended Atlantis T3 and Primesep B2 columns for additional testing.... How did the chromatography look like with buffer at pH 3 – 4 and higher? Also, was the buffer concentration the same – both at low pH (i.e. 2.3) and at higher pH values?
No, data collection is started immediatelly after start. Maybe used analytical column is well packed. As I can say from my experience, dead volume of 250x4.6;5um columns is usually about 1.8 - 2 mins.
... With the described column (250x4.6mm) and 1 mL/min flow rate, it shouldn’t be possible to see peaks eluting before 2.7 min (approx.) – dead volume. ... Do you start collecting data a minute or so after injection?
Hi Hans,The hydrogen bonding is als swamped, apparently