Noise of a detector is affected by multiple elements. One is the path length. If you increase the path length by a factor of 2.5, wouldn't you expect that in low light conditions the noise also increases by some factor around the same value.
When we compare the noise of different detectors, the sampling rate plays a giant role. The standard sampling rate on an HPLC detector maybe around 1 datapoint per second, on the UPLC, you can go to 80 datapoitns per second. It appears that you are measuring on the UPLC with a rate of 20 datapoints per second. A rather standard HPLC rate is 1 datapoint per second. This would result in a roughly 5 fold difference in noise.
If both of these things are true, I get an order of magnitude difference in noise...