ym3142 » Thu Mar 08, 2007 5:02 am
It is a long discussion.
I want to take this chance to thank all the participants, especially Uwe, Danko, Peter, Hans, and Mark. You guy’s expertise helps us learn more about the LC, both physical and chemical sides.
(only repeat what people already described) It seems that it is dominantly believed that: 1) when pH=pka, 50% is anion and 50% is neutral acid; 2) when there is enough buffer, the formation of these two forms will not cause peak to shoulder/split; 3) when there is not enough buffer, it is possible that two layers of mobile phases(these two are different) are formed in the column and sustained through the column to give peak shoulder/splitting; 4) the reason for a column to be unable to see the separation of anion/neutral acid is that proton exchange process is relatively much faster.5) further implication, that regardless the preparation of buffer, if the analyte is dissolved in a the mobile phase, there will not be any shoulder/splitting due to buffer capacity;
To my understanding, the above conclusions do not conflict from one to another and are theoretically correct. Though I do not agree with Uwe’s UV theory for supporting above conclusions. Because we know putting sodiumn benzoate and benzoic acid to any sodium acetate buffer will always give the same thing such as pH 3.2 ,both sodiumn benzoate and benzoic acid give 90% the acid, 10% the anion; 4.2 , 50% acid, 50% anion; pH 5.2 10% acid and 90% anion. In other words, what UV detector sees may have nothing to do with what happens inside a column.
Thanks