Posted: Thu Oct 12, 2006 4:44 pm
Variance should be "equal" through the whole calibration line. But as yw3142 wrote, if the error is the same, the %error for the 20µg/L sample is far more than for the 800µg/L sample. So I would suggest to use the most accurate standard as possible when using single point calibration.
But before using single point, it should be pointed out that single point is acceptable, during a validation.
Bart