5-point Calibration curve

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

35 posts Page 3 of 3

Variance should be "equal" through the whole calibration line. But as yw3142 wrote, if the error is the same, the %error for the 20µg/L sample is far more than for the 800µg/L sample. So I would suggest to use the most accurate standard as possible when using single point calibration.

But before using single point, it should be pointed out that single point is acceptable, during a validation.


Bart

I think some clarity should be made in this thread, since 2 totally different areas, and ways of calibrating, are overlapping here.

On one side there are posts dealing with typical QC/QA issues (range close to 100%, impurities etc.) which are regulated by a set of guidelines issued by ICH (I guess, it is not my field).

On the other hand there are posts like those of LewisC who speaks about analysis of substances in biological media (plasma, urine etc) which are performed according to the FDA guidance for industry on bioanalytical method validation (http://www.fda.gov/cder/guidance/4252fnl.pdf), which states that multi-level calibration samples (75%, or minimum 6, valid) and QC samples at minimum 3 levels should be used during routine analysis.

In this respect, I would suggest LewisC to read again the above guidance, since what he reported in his posts is absolutely not in line with the commonly accepted practices in the field and casts some doubts on the precision and accuracy of the concentrations he finds.

Guenzia,

All our assays are fully validated according to FDA guidelines, I was just wondering in what respect you think these guidelines have not been followed? We use a seven point calibration to cover the expected range of results, with three independently prepared QC samples at 'low', 'medium' and 'high' concentrations. The back calculated concentration of the high QC, when using our lowest calibration standard as a single calibrator, is not within ±15%, and so cannot be accepted for routine use. If anything, the FDA guidelines completely rule out the use of single calibrators.

LewisC

LewisC,

you wrote in a previous message about recovery ranging 60%-110%.
Is this accepted by FDA?

Ofcourse I'm not in your place, so I think from my place, which is definetely another working area, so don't take my messages personal :wink:

Bart

Bart,

The concentration I found for the high IQC using single-point calibration (when the single calibrant was near LLOQ) did range from 60-110% of the nominal value. This is not accepted by FDA, and so cannot be accepted. Using multipoint calibration, all three IQC concentrations fall within FDA recommended values for routine use.

I think it really is a case of different requirements for different purposes, but it is really interesting to hear different perspectives on this issue!

LewisC
35 posts Page 3 of 3

Who is online

In total there are 3 users online :: 2 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry