I am not disputing the validity of determining loss on drying by weighing the sample before and after drying. In the loss on drying procedures the repeatability of the loss measurement depends on the repeatability of the balance in the range between container + wet sample and container + dry sample. As long as the repeatabiliy of the balance is a suitably small fraction of both the weight of the sample and the loss, then the the loss measurement will also be suitably repeatable.
Within the working range of the balance, providing the conditions above are met, the actual weight of the container and the sample willl make no difference.
The critical question in this whole discussion is; "what is suitable repeatability ?". The USP has defined criteria, and JM has gone to the trouble on our behalf (thanks JM) of experimentally verifying that pre-loading a balance that has a 3 mg minimum limit still does not allow it to weigh 2 mg with a repeatability that satisfies the USP criteria. The validity of the USP criteria are not relevent to this discussion.
We all need to keep in mind that every weighing we do is a difference, because there is always a pan on the balance.
On your proposal for determining small quantities of metabolite by spiking with standard. The scheme that you propose resembles calibration by known additions ( except that this would usually involve spiking with five different levels of standard) although I have never heard of it being applied in this way. Whether or not it could work depends on what it is that determines how small the LOQ is. If the calibration is tightly linear to well below the LOQ and the LOQ is limited by random variability in individual measurements, then spiking will not help because the variability will remain too large in relation to the difference in signal that you are trying to measure. If there is no other way of adressing the problem you can force a random variability LOQ lower by running multiple replicates of the analysis and working with the means of the replicates.
If your measurements are very repeatable and the limit on the LOQ is set by the calibration deviating from linearity at low levels of analyte, then your scheme could in principal succeed by bringing the signal up onto the linear portion of the calibration. I would suggest though that it would be far better to fit a curve to the low level calibration data (even if you have to simply interpolate between the points). You should, of course, always have a calibration that spans the concentrations of analyte in your samples.
Have you ever actually done an analysis in the way that you propose ? Were you able to validate it in terms of accuracy and precision ?
Peter