Back Pressure Regulator Problem
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
41 posts
Page 3 of 3
Ok, so you're saying that there is a leak based on the fact the baseline is smooth, but what exactly am I looking for?
-
- Posts: 2846
- Joined: Mon Aug 30, 2004 7:17 am
What I said is that if by some single action the baseline became smooth the fault was caused by what you corrected with this action (except if the problem is corrected spontaneously just at that time).
You are looking for faults in the plumbing.
I still wonder what spectra look like of different portions of this "baseline".
You are looking for faults in the plumbing.
I still wonder what spectra look like of different portions of this "baseline".
-
- Posts: 25
- Joined: Fri Feb 03, 2006 6:22 pm
I could some pics of the spectra, if you'd like to see them.
-
- Posts: 2846
- Joined: Mon Aug 30, 2004 7:17 am
If you can´t imbed them here you could explain them. In my experience air peaks have produced specs which just rapidly come down from 200nm, staying higher to at least 300nm (didn´t go above) than a "better" baseline. Compounds with a maximum above 200 nm show that maximum, of course.
-
- Posts: 25
- Joined: Fri Feb 03, 2006 6:22 pm
I will place the spectra here if I can. I'm not sure I could explain the spectra satisfactorily.
-
- Posts: 25
- Joined: Fri Feb 03, 2006 6:22 pm
Here is the chromatogram:
-
- Posts: 2846
- Joined: Mon Aug 30, 2004 7:17 am
Just looks as if you didn´t leave the system alone, injecting stuff, including air, adjusting the baseline.... instead of letting it come to equilibrium before assessing it (in other words, this discussion has dragged on, apparently because you didn´t really tell what you did). What about the UV specs?
-
- Posts: 25
- Joined: Fri Feb 03, 2006 6:22 pm
The UV specs are flat baselines, except for the injection peaks.
The other spectra show a baseline that is elevating, then dropping when I've done nothing to it and let it sit. I was very careful not to inject air, but that doesn't mean I didn't.
My biggest problem right now is inexperience. My apologies for letting this drag on as long as it has. In the future I'll try not to inundate everyone with problems that I should be solving myself.
Thanks to everyone who has helped me, especially Mr. Mueller.
The other spectra show a baseline that is elevating, then dropping when I've done nothing to it and let it sit. I was very careful not to inject air, but that doesn't mean I didn't.
My biggest problem right now is inexperience. My apologies for letting this drag on as long as it has. In the future I'll try not to inundate everyone with problems that I should be solving myself.
Thanks to everyone who has helped me, especially Mr. Mueller.
-
- Posts: 2846
- Joined: Mon Aug 30, 2004 7:17 am
Just play around a bit more, you will get the hang of it.
-
- Posts: 1
- Joined: Fri Jun 09, 2006 3:49 pm
Hi Steve,
I was wondering, you mentioned you had the U-610 BPR from Upchurch... that one is actually rated to 750 psi. Did you actually have the U-605 (I hope) which is rated to 40 psi? You'll know by the label on the body of the BPR or by the colors of the cartridges. The 750 psi cartridge has a black body and blue end cap, the 40 psi cartridge has a tan body and blue end cap. If you did have the 40 psi cartridge inline and the pressure was much higher, it sounds like a defective part. You might want to contact Upchurch's customer service department about returning the part.
Thanks,
Caitlin
Upchurch Scientific
R&D Chemist
I was wondering, you mentioned you had the U-610 BPR from Upchurch... that one is actually rated to 750 psi. Did you actually have the U-605 (I hope) which is rated to 40 psi? You'll know by the label on the body of the BPR or by the colors of the cartridges. The 750 psi cartridge has a black body and blue end cap, the 40 psi cartridge has a tan body and blue end cap. If you did have the 40 psi cartridge inline and the pressure was much higher, it sounds like a defective part. You might want to contact Upchurch's customer service department about returning the part.
Thanks,
Caitlin
Upchurch Scientific
R&D Chemist
-
- Posts: 25
- Joined: Fri Feb 03, 2006 6:22 pm
Caitlin,
It was the black body with the blue cap. I thought I was ordering the 40 psi cartridge, but it seems I didn't. The part number was given to me by someone else who had a similar one, but she had the 40 psi cartridge.
Steve
It was the black body with the blue cap. I thought I was ordering the 40 psi cartridge, but it seems I didn't. The part number was given to me by someone else who had a similar one, but she had the 40 psi cartridge.
Steve
41 posts
Page 3 of 3
Who is online
In total there are 28 users online :: 4 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Amazon [Bot], Google [Bot], Majestic-12 [Bot] and 24 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Amazon [Bot], Google [Bot], Majestic-12 [Bot] and 24 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science