peak tailing and sensitive lose may caused by contamination. what i do is cut the column (should >50cm) and change inlet , split pan (glod pan) may be also need change if you do trace analysis.
Advertisement
Discussions about GC and other "gas phase" separation techniques.
Can you show us some chromatograms over the full retention time range? Preferably a good one and one with the problem present. Do you see this contamination in SIM? In standards and samples or only samples?guess what, the contamination came back... Honestly I dont know what to do right now, peaks are still in good shape. I will measure some samples and see how bad the tailing gets. Although I dont have any hopes since I aint using a guard column at the moment ;(
Don't bother running the samples, the performance will not be stable if you have major contamination.guess what, the contamination came back... Honestly I dont know what to do right now, peaks are still in good shape. I will measure some samples and see how bad the tailing gets. Although I dont have any hopes since I aint using a guard column at the moment ;(
If you happened to get one bad tank of gas it could have contaminated the lines and is now slowly bleeding through the system. I have had that happen before and it is a real pain to get cleaned out.I am still working on that problem, I did what Peter suggested although I aint really sure what is the outcome.
I did some blank injections, it did decrease VERY little over 10 injections. I also saw already that the first injection is a lot more contaminated than the following here. I was able to verify that indeed after letting the instrument sit for a few hours I get a higher contamination peak. Then I did the 0-Volumen injection (which was a 0.1µL air injection). However the signal was a little increased, it was not much, I would say little higher than the blank runs.
Since the contamination decreased now by a factor of 10 since 2 weeks I suspect that it is indeed from the instrument, being slowly evaporated into the column over time. To be honest the only remaining part I did not clean/change is some steel tubing now. I could heat it up to see if the contamination increases.