Posted: Thu Jan 26, 2006 1:22 am
Syx, what do they call the cis and trans isomers. At what point in the molecule is the cis-trans transformation? I am used to cis-trans around a double bond, but I can't see it here?

The cis-trans position of methyl and H in cyclohexyl ring (upper right). Cis-isomer is process impurity, not a degradate.
Here are some articles related to Glimepiride isomers and the used condition for separation:... if you are referring that article over isomer determination in crude samples of glimepiride, the one I send to you, thats RP HPLC.
Song Y, et al. J.Sep.Sci. 2003, 26, 1595-1597
Column: Dikmonsil C18 5um 250 x 4.6 mm
Mobile phase: Methanol:acetonitrile:HOAc-NH4Ac buffer (1.5M
Flow rate: 1.0 mL/min
Injection volume: 10 uL (sample concentration 100 ppm)
Result: Rt of trans-isomer = 23.018 min; cis-isomer = 24.885 min
Jun W, et al. Chinese Journal of Spectroscopy Laboratory 2004, 21(1) 150-152
Column: Zorbax SB-C18 (250 x 4.6 mm, 5um)
Mobile phase: Methanol: HOAc-NH4Ac buffer (0.05M, pH 3.
Flow rate: 1.0 mL/min
Temperature: 20 C
Result: Rt of trans-isomer = 22.4 min; cis-isomer = 23.7 min
Khan MA, et al. J.Pharm.Biomed.Anal. 39(2005) 928-943
Column: Phenomenex Luna C8(2) 5 um, 250 x 4.6 mm
Mobile phase: Phosphate buffer (pH 7.0):acetonitrile:tetrahydrofuran = 73:18:9
Flow rate: 1.0 mL/min
Temperature: 35 C
Result: Rt of trans-isomer = 43.275 min; cis-isomer = 41.383 min
USP29–NF24 Page 1001
Column: 3-mm × 15-cm column containing 5-µm packing L20
Mobile phase: Transfer 100 mL of isopropyl alcohol into a 1-L volumetric flask, add 1 mL of glacial acetic acid, dilute with hexane to volume, filter, and degas.
Flow rate: 0.5 mL/min. [ NOTE— The analyses could also be performed with 4.6-mm × 15-cm, 4.6-mm × 25-cm, 4-mm ×12.5-cm, or 4-mm × 25-cm columns containing packing L20. It is recommended that the flow rate be adjusted to about 1.1 mL/min for a 4.6-mm column and to about 0.8 mL/min for a 4.0-mm column.]
The relative retention times are 1.0 for glimepiride and not more than 0.9 for the cis-isomer, and the signal-to-noise ratio of the cis-isomer peak is not less than 15.
See? Need a lot of time to separate the isomers.
Btw, have you any reference about substances (excipients, especially in tablets) those could adsorb amlodipine? I have difficulty to get good recovery. I always got below 98.0% though I avoid filtration and use many types of solvent.the compound whose chromatograms i posted in the very begining was Amlodipine Besylate but we have observed this with Ramipril.