Re: musty odor in drinking water
Posted: Fri May 20, 2016 10:15 am
You can do a PTV with the methanol as well - start just below the boiling point and ramp quickly to 240C or so (I sometimes do this since it seems to lead to less crud in the inlet). Start with a 1 ul injection and no solvent purge (in effect a splitless injection) and increase the volume with a solvent purge if you think that you need more sensitivity.Reading through the question and responses, is there a reason you couldn't manually sample the headspace with a syringe and run a splitless injection?
We've already done that but we couldn't find any peaks different from the blank.
The MeOH extract has a musty odor so my unknown is in it.
I'll try some different techniques to get it analysed. I have very little experience with SSL liq injections (I mainly do 50µl CH2Cl2 PTV injections) but I'll have a go.
thanks again for all the input it is much appreciated!!!