GC-FID active sites Jet?

Discussions about GC and other "gas phase" separation techniques.

34 posts Page 3 of 3
Yea I tried that on some old columns that were tailing badly. I used the solvent pump to fill them with DMCS and TMCS in chloroform and plugged the column with old septa and kept them in an oven at 55 deg overnight then reconditioned them in the morning. It made little difference. I also tried injecting the reagents splitless over and over.

I doubt it is stationary phase damage in this case as I usually see almost all peaks except straight hydrocarbons tailing when there is stationary phase damage.
I just ran a standard and it looks a million times better. Better than the GC/MS. There must have been some sort of contaminant in the GC column.

Thanks for all the help.
I just ran a standard and it looks a million times better. Better than the GC/MS. There must have been some sort of contaminant in the GC column.

Thanks for all the help.
I confess myself amazed ! Thanks for the feedback.

Peter
Peter Apps
Glad that you have solved what I earlier described as "your 3 pipe" problem
:D
Regards

Ralph
Regards

Ralph
34 posts Page 3 of 3

Who is online

In total there are 78 users online :: 0 registered, 0 hidden and 78 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 78 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry