Re: New to Chromatography - Manual Injection Technique

Posted: Fri May 10, 2013 7:18 am
by Peter Apps
Hi Michael

In HPLC you have a mechanical pump with wearing parts, in GC you have a gas cylinder. In HPLC you have to make up mobile phase every couple of days, in GC you change a cylinder once a month. In HPLC there are thousands of mobile phase recipes, in GC there are three (and they only affect the speed of the analysis). In GC you have a temperature programme, in HPLC you have mobile phase gradients. There is no HPLC equivalent of an FID (which is both a good and a bad thing). Choosing the right column is more critical in HPLC.

Peter

Re: New to Chromatography - Manual Injection Technique

Posted: Thu Jun 06, 2013 5:21 pm
by MichaelStuart
I'm happy to report that things seem to be going swimmingly. After going through three controllers, four towers, and two trays, I finally have a working auto-sampler. I think I was also saturating the column. I adjusted my split ratio from 10:1 to 20:1, and am now getting consistently about a 1%-2% deviation between determinations. I can live with that.

It seems the fast injection of the auto-sampler has circumvented the inlet discrimination that I was noticing.

Peter, thanks again for your help. :wink:

Mike

Re: New to Chromatography - Manual Injection Technique

Posted: Fri Jun 07, 2013 7:51 am
by Peter Apps
A pleasure.

Thanks for the feedback and congratulations on getting it sorted.

Peter

Re: New to Chromatography - Manual Injection Technique

Posted: Fri Jun 14, 2013 6:13 am
by MichaelStuart
I feel like it's kind of an off the wall question.. However, Is it possible to qualify/quantify lead in a paint chip with my 5890-s II FID/ECD?

Re: New to Chromatography - Manual Injection Technique

Posted: Fri Jun 14, 2013 7:01 am
by Peter Apps
I feel like it's kind of an off the wall question.. However, Is it possible to qualify/quantify lead in a paint chip with my 5890-s II FID/ECD?
Possible yes, probably with some pretty exacting sample prep aiming to make a volatile derivative of the lead such as lead tetraethyl.

Practical - definitely not.

Peter

Re: New to Chromatography - Manual Injection Technique

Posted: Fri Jun 14, 2013 12:03 pm
by chromatographer1
Gc is definitely not the preferred methodology.

Atomic absorption analysis is the preferred methodology for trace levels, not necessarily a test for lead paint. Wet chemical analysis heavy metals is a nice quick and easy test.

best wishes,

Rod