This is strted as a $ sving project, nitrogn carrier while GCs were idling/not in use. Still validating some of the ISO 9000 QA methods but have reduced He usage 80%.
Advertisement
Discussions about GC and other "gas phase" separation techniques.
For the same peak capacity, hydrogen is about 70% faster than helium in GC-MS applications and in high peak capacity applications. Hydrogen is about 25% faster than helium in low peak capacity non-GC-MS applications. And higher speed typically comes with lower (better detection limits). This suggests that, if shorter analysis time and/or lower detection limit are important, hydrogen carrier gas is the best choice.Argon may be a replacement for some inert gas applications since it is readily available as almost 1% of the atmosphere.
James, I can't see a point of doubling hydrogen supply system for emergency situations like power outages. Are your GCs battery operated ?... We also have hydrogen tanks at the other end of our supply lines outside just in case of power outages like when we were out for 8 days after an ice storm a few years ago. Easy to set up if you set the tank regulator a few psi lower than what the generator is running with a back flow preventer in line it will kick in once the generator stops.
I don't know about James, but I have battery operated GCs - the batteries are in a 6KVA UPS.James, I can't see a point of doubling hydrogen supply system for emergency situations like power outages. Are your GCs battery operated ?... We also have hydrogen tanks at the other end of our supply lines outside just in case of power outages like when we were out for 8 days after an ice storm a few years ago. Easy to set up if you set the tank regulator a few psi lower than what the generator is running with a back flow preventer in line it will kick in once the generator stops.
We have UPS on a few but mostly it is to keep a flow for the older ones with mechanical flow controllers if power interrupts the generator. The generator has a power backup but it will only last a few minutes, if the power is off overnight then it is better to have some flow through those columns than non at all.James, I can't see a point of doubling hydrogen supply system for emergency situations like power outages. Are your GCs battery operated ?... We also have hydrogen tanks at the other end of our supply lines outside just in case of power outages like when we were out for 8 days after an ice storm a few years ago. Easy to set up if you set the tank regulator a few psi lower than what the generator is running with a back flow preventer in line it will kick in once the generator stops.
What is the hold up time of this UPS in your typical load configuration ?...
I don't know about James, but I have battery operated GCs - the batteries are in a 6KVA UPS.
Peter
Hi Rod
I found a GC battery power backup of 6000 VA fascinating.What a lab that is !
best wishes,
Rod