hi again
i have tried running my sample again for the last time.. i've given up already..
we'll be sourcing another API supplier to confirm if the API has the problem..
but i have some fews questions though..
during this last trial, i have replaced the wash solution with 100% acetonitrile, i've cleaned and sonicated the port filters, and ive changed the line that im using for the gradient (before i was using line C and Line D for my MPA and MPB) now i have used Line A and B for the mobile phases.. before i start the run i prime the wash solvent and the needle wash.. i run 100% acetonitrle in my column (symmetry shield 150 x 4.6 ) for 15 minutes then i equilibrated the column with 5% acn ( i just run 5% acn for 30 minutes)...
(note : i have used the BP method during this run.)
this is the chromatogram for my thiamazole :
the unknown peak that i have seen before is not seen in this chromatogram now..
there is no unknown peak now...
then i have run the carbimazole API only but i still got the unknown peak.. (i have run this one without the reference wavelength.. the one that i run in the agilent has the reference wavelength at 360nm, and i think this one made the unknown peak disappear, but i cant use this reference wavelength because in this wavelength the carbimazole peak is also detected., )
also with my sst.. its all the same for carbimazole..

..
i really dont know what to do then i tried do the test for api RELATED SUBSTANCES using the bp.. with this conditions...
Column:
— size: l = 0.15 m, Ø = 3.9 mm,
— stationary phase: octadecylsilyl silica gel for chromatography R (5 µm).
i used the waters symmetry shield for my column..
Mobile phase: acetonitrile R, water R (10:90 V/V).
detection : 254 nm
injection volume :10 ul
Flow rate: 1 mL/min.
5 mg the sample (carbimazole API was dissolved and diluted in 10 ml of 20% ACN)
i have no other impurities in my carbimazole api. and i have only 1 peak...
then i tried running the same sample i used in my gradient run in isocratic condition using 20% ACN MPB..
i still get 1 peak for my carbimazole, and a thiamazole peak..
why do i have an unknown peak beside my carbimazole peak during the gradient run, but when i run the same sample in isocratic ung my MPB (20% acn) i only got i clear peak for carbimazole..
can anyone help me..
i still got time until the new batch of API arrive..
thank you so much!!