found old applications for adenosine in RP/cation-exchange mode. You can see how by changing strength of ionic intercation you can increase retention time:
Image

You can also increase retention by reducing amount of ACN from 50% to 5-10%. Retention is controlled by both amount of ACN (RP) and amount of ions/pH (cation-exchange)
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
Oooh thank you Vlad! Thank you very much for your post.

Then... to observe this image carry me to think that the best compromise to peak shape and retention of adenosine is give from the use of Primesep 200 and acetic acid. It's right?
Another thing that I observe is that in all test you use 50% of MeCN in mobile phase. It's high as concentration. In my test if I use more than 10% of it Inosine is poor or not reteined really. This difference is by from cation exchange mode? Can I reproduce this condition on my cloumn/system? With this thing I would can to work in isocratic mode!

I reply to you Danko. I'm the best crhomatographer in my lab. I know that I'm unprepared a lot on many discussion. But nobody born masterly, or not? ;) I must study more too. I know it!
About column... When I buy it I think that was the best choice. But you and other guys had demonstrate the opposite. Ok. I'm sorry for my error, but I can't go behind to repare to it. I must found the best solution with PFP kinetex column. If you want to help in this my mission, you're welcome :)

Thanks!
Considering everything in your mixture (Isoprinosine or Methisoprinol), I would go with 5-10% ACN and TFA in the mobile phase on Primesep 100 column. This will allow you to retain and separate inosine, acetamidobenzoic acid and dimethylaminoisopropanol. You will need ELSD for dimethylaminoisopropanol.
Inosine will retain by weak reversed-phase and cation-exchanage mechnaisms
acetamidobenzoic acid will retain by reversed-phase mechanism (at low pH created by TFA it is not inized)
dimethylaminoisopropanol will retain by weak reversed-phase and strong cation-exchange mechanisms.

Unfortunately this is not possible on the column you have, but I think it is better to change column now and redevelop the method now with the goal of separating all components of your formulation.
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
Inosine is very polar and there is little you can do to increase its retention in reversed phase mode. Ion paring will not be effective since it is difficult to ionize inosine (requires pH ~1 or pH ~10).

The best you can do with RPLC will be to use gradient elution starting at 100% water with 0.1% phosphoric acid (channel A) and 0% methanol with 0.1% phosphoric acid (channel B). Run 2 blank gradients from 0-50% B prior to running your sample (make 5 injections of sample to check reproducibility).

The column must be re-equilibrated with 5 column volumes of 100% A prior to the next gradient run. Run each analyte separately and post your chromatograms here.
A. Carl Sanchez
Inosine does not require pH 1 or 10 to be retained by ion-exchange mechanism. How would you explain retention on mixed-mode if it is not ionized? As you can see from my chromatograms you have cation-exchange even at pH 3 and 4 on mixed-mode cation-exchange column with plenty of retention (K'>20 in some chromatograms). Less ions, higher pH - more retention. Mixed-mode basically mimics ion-pairing chromatography. Also in mixed-mode you have synergy of two mechanism.
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
If what you say about inosine's ionization in pH 3 and 4 solution is true (which I disgaree with) then this user should have been succesful with the early recommendation by Mattias to use an anionic/acidic ion pairing agent (heptane sulfonate) with acidic mobile phase ~pH2 (0.1% TFA).

The ion pairing agent (heptane sulfonate) did not increase retention.

In the meantime hopefully almegot will try gradient elution starting at 100% aqueous with his RPLC column.

The sample should be dissolved in water (no organic)
A. Carl Sanchez
Please take my explanation with a grain of salt, and don't consider this as another pitch to promote our columns:
I think that when you have two interactions you enhance each of interactions by having a second one, that is why for example if you have hydrophilic molecule with weak basic group you still can control retention by amount of ACN, pH and amount of buffer. You can run hydrophilic basic compound on RP, and have it close to void even at very low organic, when you switch to RP/cation-exchange you will see effect of ACN even going from 30% to 20%. How would you explain drastic increase of retention for adenosine on chromatograms above, when we simply changed pH and amount of ions in the mobile phase? retention increased several folds for the strongest column.
You are saying that you need to be at pH 1 or 10 to have ion exchange and my example says that you don't.

In mixed-mode chromatography 1+1 is never 2, it can be 3, 4 or any other number (I am talking about strength of interactions)
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
The pKa of the most basic N of adenosine is ~4 while the pKa of the most basic N of inosine is ~1.

Therefore I believe it is very unlikely that the retention on a mixed mode (RP/cation exchange) column will be any different from what was seen with heptane sulfonate.

A strong anion exchange column has a better chance of working than a cation exchanger.
A. Carl Sanchez
If the most basic pKa is 4 for adenosine, retention at pH 4 should be less than at pH 3 and this is not the case.
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
I can't enter in your discussion. I'm not prepared on these points.

I don't know what is the best choice for my analysis. HILIC, cation-exchange, ecc. But one thing I know for certain. I must use the kinetex PFP column. Right or wrong that is it. I must create a best method on this column. :)

Then I write my considerations about it. On the chromatograms that you post Vlad I see that Inosine have a peak shape similar to my shape on kinetex.
I'd open this post to search a way to increase the Inosine peak height at the first time. The reason is that In my sample Inosine and 4-acetamidobenz are mixed in 1:3 ratio. Then the 4-acet. peak go out of my detector full scale, and Inosine no. I'd hope to create a similar respound to adjust Injection/concentration of my sample solution for the best compromise for the peak height.
To increase the less retention problem is the second thing.
To conluse my discourse. I think that there is not possibility to arrange the 2 componente respound with column change. What you think?

Another question. Can I have it with change detector wavelenght?

About retention problem. Ok With cationexchange column Inosine has good retention. But I don't have it :)
I must work in other direction. If there is not possibility to improve retention on my clomun I'll work with only water with low pH value (for 4-ac.benz. peak shape and retention) and I use a gradient way. What I can do to reduce classic gradient ghost peak? To use a very smal (5%) MeCN from the start of analysis?

Thanks!
Where this is method heading? Are you going to validate it and use in the future? I ordered inosine and will at this in our lab, once I get some results I can probably send you a demo column.
Vlad Orlovsky
HELIX Chromatography
My opinions might be bias, but I have about 1000 examples to support them. Check our website for new science and applications
www.helixchrom.com
Where this is method heading? Are you going to validate it and use in the future? I ordered inosine and will at this in our lab, once I get some results I can probably send you a demo column.
Yes. I'll use this method in future.
Thank you Vlad. If you want to try it, try it! And for demo column... we'll speak about it in future! Thank you.

meantime... in my other post one user write that Kinetex columns have a very great number of free silaniols. This information is correct? If it's correct they can distrube Inosine eluition, right?
Maybe it’s a good mixed mode column (cation exchange + reversed phase)?
:lol:
Best Regards
Learn Innovate and Share

Dancho Dikov
I must use the kinetex PFP column. Right or wrong that is it. I must create a best method on this column. :)

I'd open this post to search a way to increase the Inosine peak height at the first time. The reason is that In my sample Inosine and 4-acetamidobenz are mixed in 1:3 ratio. Then the 4-acet. peak go out of my detector full scale, and Inosine no. I'd hope to create a similar respound to adjust Injection/concentration of my sample solution for the best compromise for the peak height.
To increase the less retention problem is the second thing.
To conluse my discourse. I think that there is not possibility to arrange the 2 componente respound with column change. What you think?

Another question. Can I have it with change detector wavelenght?
Thanks!
The UV max for inosine is ~255nm and the UV max for 4-acetamidobenzoic acid is ~270nm.

Using 255-260nm should give the maximum inosine signal.
A. Carl Sanchez
alemaggot
can you please explain what was the reason you chose to try a PFP column.
the fact is that it is a poor choice for your case; but more importantly i want to understand the reason why?
learning what got you to choose it, I think is an important lesson here that can help least experience analysts.

thank you in advance
82 posts Page 3 of 6

Who is online

In total there are 80 users online :: 2 registered, 0 hidden and 78 guests (based on users active over the past 5 minutes)
Most users ever online was 11462 on Mon Dec 08, 2025 9:32 pm

Users browsing this forum: Ahrefs [Bot], Semrush [Bot] and 78 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry