lmb » Thu Apr 14, 2011 10:26 pm
													
 
					
						I think that, under ideal conditions, 3 is the correct answer because the carrier gas is a compressible fluid. Otherwise (in LC, for example) the correct answer would be 1. These answers do not depend on the lengths of the two sections of the column.
It looks like the problem is mathematically simple, but logically it is a bit trickier. Here is why I think so.
Due to the gas decompression along the column, the gas velocity increases in the direction from the column inlet to the column outlet. The column consists of a coated and a bare segment. In configuration A, the bare segment is on the inlet side and the coated segment is on the outlet side. Let tM_bare_A and tM_coated_A be the hold-up times of these segments in configuration A. In configuration B, the column is reversed so that the bare segment gets on the outlet side and its hold-up time (tM_bare_B) becomes smaller. Conversely, the coated segment gets on the inlet side and its hold-up time (tM_coated_B) becomes larger. Reversing the column increases the time of the residence of a compound in one segment while reducing the time of that compound residence in another segment. This complicates a purely logical approach to the problem (i.e. solving it without the exact mathematical evaluations).
Now is the time to bring the retention to the picture. If k is the retention factor in any segment of a column then parameter (1+k) can be viewed as a magnifier of that segment's hold-up time (tM). Thus, the retention time in a segment is equal to (1+k)tM where tM is that segment's hold-up time. The compounds are retained in the coated segment – the one whose hold-up time in configuration B is larger than in configuration A. Now, it is important to recognize that the net hold-up time in configuration B is the same as the net hold-up time in configuration A. The net retention time in a column is the sum of the hold-up time in one segment (in the bare segment to be precise) and the magnified hold-up time in another segment (in the coated segment to be precise).
For any retained peak, the net retention time in configuration B is larger than it is in configuration A because the reversal of the column direction 
a) does not affect the net hold-up time (the sum of the hold-up times in the coated and the bare segments),
b) magnifies by the same factor of (1+k) the hold-up time that is larger in configuration B than it is in configuration A.
Here is a more formal expression of the same logic.
If tR_A and tR_B are the net retention times in configurations A and B, respectively, then:
tR_A = tM_bare_A + (1+k)tM_coated_A
tR_B = tM_bare_B + (1+k)tM_coated_B
tR_B = tR_A when k = 0
tM_coated_B > tM_coated_A
Therefore, tR_B > tR_A when k > 0.