The oxygen level on the other GC is 2.6%With a mass spectrometer - you will always see some oxygen. There is always a small leak through the seals of the mass spectrometer. An additional source of air entry into at GC/MS system can be the split/splitless inlet. If your head pressure is low enough, air will diffuse into the inlet if the split vent is open. This can be decreased by increasing the split flow. And you can have air enter through the septum purge and through a cored septum.
You should see some air background on the other GC system. How does it compare when you run the leak check on that instrument?
The reason I suggested injecting methanol was to eliminate reaction of anything but methanol from the picture. If the contaminant peak is still there, it has nothing to do with analytes or other contaminants. If the strange peak is still there, the next step is to push the run button on the GC as if making a manual injection. This will eliminate anything coming in through a syringe or in the solvent.
If you look in methanol for formaldehyde, you will find some. If you check the specifications on your methanol, you will most likely find that it is certified to contain less than some level of formaldehyde. And that specification is good on the date when the methanol was analyzed. If your methanol has been sitting in a stockroom for a couple of years, the level of formaldehyde can increase. Also expect to see some formic acid and methyl formate.
If the methanol has been stored in a clear glass bottle in the lab, look for these to grow even more.
Alle bottles are HPLC gade and less than 6 months old and they are stored in brown bottles...
Ill check the specs for formaldehyd ect.