If you look at the Amide application note book you will see that when they use 0.2% TEA in the mobile phase they run at 35C. When they run at high tem they run at much lower concentration of TEA... I always suspected that was because of column life issues at h igh temp at high pH

I suspect that is what is happening with your columns
Jamie,

You have two concerns here to be worried about:

1) Effect of NH4OH on the column material vs. the effect of TEA. Presumably you will be in a position to report on that soon.

2) Reaction of the additive with your analytes. TEA is a tertiary amine and so cannot form a Schiff base. The ammonium ion in NH4OH may well form them with your reducing sugars. That's why nobody uses it to increase the pH (and thus increase the rate of mutarotation).
PolyLC Inc.
(410) 992-5400
aalpert@polylc.com
32 posts Page 3 of 3

Who is online

In total there are 29 users online :: 4 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Amazon [Bot], sara32, Semrush [Bot] and 25 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry