Hi Everyone.
I've followed this topic for a while now with interest and I thought I'd add my experience to the mix.
I've known about this issue for close to 10 years now and it has been extremely frustrating. Firstly, it's weird and hard to diagnose. Secondly, because of the fact that it's weird, some other analysts may not believe it's a real issue and it can take some convincing to prove that it's the machine that's the problem and not something wrong with the analytical technique.
To summarise my experiences:
I analyse canola, sunflower and sometimes, camelina seed for fatty acid composition using area normalisation. These three types of seed have a number of fatty acids that range from extremely small up to huge for certain ones like C18:1. I use SGE BPX-70 columns.
I previously worked in a government lab where they had stopped using the GC-2010 and were sticking with the older GC-17A. I was told that the GC-2010 didn't provide accurate results and in AOF interlaboratory proficiency trials, the results from the GC-2010 would have made our lab look bad. I found this to be the case in my own tests. Basically, large peaks were too large, small peaks were too small and some mid-range peaks barely changed relative to the other fatty acids. So for example, from memory, if C18:1 was 65% on the GC-17A, it came out at around 67.5% on the GC-2010. C18:0 may have been 2% on the GC-17A but 1.6% on the GC-2010. The AOF interlaboratory results agreed with the GC-17A but the GC-2010 results would have been outliers in many cases. Also, I ran AOCS Standard Mix 6A on both systems. This AOCS standard agreed with the GC-17A but the GC-2010 was way out (>2 standard deviations from the mean for various fatty acids).
I contacted Shimadzu about it back then and for many months, while discussing this with one of their engineers, I tried all kinds of possibilities including gas flows etc as some of you have tried. In the end, nothing worked. So my manager arranged for the GC-2010 to be swapped with a GC-17A. After that, we had no more issues.
Swapping for a GC-17A is no longer feasible as this GC is no longer supported by Shimadzu and when parts run out, the GC will be history.
Fast forward to my current private lab where we have a heap of GC-2010s. When I started here, I originally wanted an Agilent machine because I didn't want the same GC-2010 issue to pop up and cause me grief. I was overruled and therefore, had to hope for the best. Luckily, the GC-2010 plus that I got gave accurate results. AOCS standard Mix 6A results were accurate. We also got a Shimadzu GC-2010plus GC-MS/FID system and it too provided good results.
However......
About a year or two ago, the FID jet got blocked on the GC-MS/FID system. We removed it, unblocked it and reinstalled it. After that, the results were bad like they were in my previous job- large peaks gave results that were too large, small peaks gave results that were too small mid range peaks were almost unchanged.
Then last year, as part of a new preventative maintenance program, the FID jet was replaced in the other GC-2010 that was working fine. After that, the results were bad.
But now it was getting clear that it was directly related to the FID jet. I asked for Shimadzu to get involved again to try to sort things out. They were able to replicate my issue in their lab using my column and samples. While they tried to sort it out, I tried a few other things. Based on comments in this forum, I tried using a 0.8mm jet rather than the 0.3mm jet. The results improved somewhat, but the sensitivity dropped too much. Plus the results were still a bit skewed.
So I asked one of the Shimadzu engineers to come in work on the issue. I asked him to bring a heap of FID jets as well. After a number of failed attempts at modifying some parameters, the FID jets were swapped in and out one by one and voila! We found one that happened to bring the results back in spec. Since then, it has worked as brilliantly as before and I've had no more issue. Noone is allowed to touch the FID jet in that GC anymore! The good new results were confirmed by AOCS Std Mix 6A and also by good fatty acid results that our lab got from a Canadian interlaboratory certification program.
I had no such luck with another GC 2010 that I had recently installed. It worked fine at first. But then the large peaks became larger, the small smaller. It wasn't as large a shift as for the other GC. But it was concerning enough as there would now be less agreement between this GC and the one that was fixed. After trying about 5 FID jets, none could be found that brought the results back. So this GC isn't being used for fatty acids at the moment.
While all this was going on, a Shimadzu analyst gave me some gas flows that brought the results back into spec. I tested them on a couple of GCs and it worked. The only down side is that the sensitivity dropped off. But for a GC that was out of spec, it brought it in spec. For a GC that was already in spec, it made virtually no difference.
The gas flows for my system are:
Makeup gas (nitrogen): 5mL/min
Air: 270mL/min
Hydrogen: 30mL/min
To improve sensitivity, I was given another set of flows:
Makeup gas (nitrogen): 20mL/min
Air: 175mL/min
Hydrogen: 40mL/min or 30mL/min
The new flows did improve sensitivity. But I've had more trouble with noise. I had to flip the FID lid closed to stop air movement disrupting the flame. These flows are still a work in progress. There may be better flow rates that improve accuracy, sensitivity and flame stability.
I've been informed that Shimadzu is bringing out a new GC model with a redesigned FID. Hopefully, this new FID works far more accurately and reliably than the current GC-2010 one. If so, it's been suggested that it might be able to be swapped directly with the one in the GC-2010 to save having to buy the new model GC. To test it properly, it needs to first of all work in the new GC. Then it needs to work in the GC-2010. Then it needs to work no matter how many times the FID jet is swapped with another. This last one is crucial because I've already experienced GCs that work fine from installation. But once the FID jet is replaced for the first time or removed and reinserted, the problem suddenly appears.
So time will tell if this issue will be resolved once and for all.
Jason