Yes, this is HILIC.is this HILIC?
I also like to get a second opinion and this is what this forum is for.sorry to have re-asked something that ought to have been obvious.
Dr. Neue, that's an interesting point to consider. Luna Amino is not standard amino propyl chemistry and does not share their bleed charateristics.Standard amino columns bleed like crazy during the initial runs, and this can be the cause of clogged outlet frits.
That did remind me of another potential issue.
Flora,
Have you flushed this column with IPA? New amino columns are shipped with normal phase solvent (hexane/IPA) which must be purged with 100% IPA (10 column volumes,CV) prior to use in HILIC mode. Please perform this purge/flush prior to doing any further work. Be sure to reduce the flow rate so as to not over pressuize the column. Inadequate purging will cause a host of different problems including unpredictable and poor chromatographic performance as well as very slow equilibration. After purging with IPA equilibrate the column with 20 CV of your initial mobile phase.
It's easy to create durability data for NH2 phase - just pump water through through the column.
Imtakt compared durability of Unison UK-Amino vs. standard NH2 phases:
http://www.imtaktusa.com/site_media/fil ... TI304E.pdf
Hi Flora,
How is your analysis working out?