Not sure that this is what's happening here, but there is a mechanism for what you describe.
DR

Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
Changes in the RI of the MP is a good reason to use a reference wavelength:Update / New Information:
We received some additional information and new data this morning that I thought I would share. I've also got a new hypothesis that I'd like feedback on.
1) The lab confirmed that all recent runs have been on Aquity H-Class UPLCs with Quaternary pumps and TUV detectors. They confirmed that they have NOT been using a reference wavelength (so that hypothesis is officially shot down...)
2) They did perform a run on H-Class system with a PDA recently (again, no reference wavelength applied), and the profile did look significantly better (almost no downward drift); however, they confirmed that one other factor was changed at the same time for this run: They didn't filter the buffer for this run (they've typically filtered the stock buffer using a 0.2µm nylon membrane). So, this provides us 2 potential contributing (and currently confounded) factors for further evaluation: Detector type and filtration. I'm interested in anyone's thoughts on whether or a nylon filter may be contributing a contaminant to the MP.
3) I'm realizing that I did not specify the column temperature in my original post (50C). One of my colleague brought up the idea that changes in refractive index of the mobile phase may be contributing to the baseline drift. More specifically, if the mobile phase is at a high and/or inconsistent temperature when it reaches the flow cell, this can cause a higher degree of sensitivity to RI changes in the MP. We're thinking that this may align with the recent observation that baseline drift was different when using a PDA detector, as the PDA design may be less susceptible to this effect (maybe the PDA flow-cell has more thermal mass, thus stabilizing the MP temp in the detector?). Does anyone have experience with RI and/or MP temperature variation affecting the baseline?
Yeah, I agree that phosphate is not the best buffer at pH 3.0. Ironically, we were forced to move away from ammonium formate to potassium phosphate because their systems were so noisy when they ran with ammonium formate (we couldn't get adequate S/N at LOQ). They spent quite a bit of time trying (unsuccessfully) to get noise down to acceptable level before making that switch. In hindsight, the buffers switch appears to have been a band-aid for a bigger issue...::reread OP::
pH 3 with phosphate - not ideal, (you're not in a pH range where the PO4 provides any buffering capacity)... so even if your stocks are perfect and clean, it could still be the MP causing problems.
What MP glassware are you using? Schott glass can be problematic with ACN and very sensitive UV detectors...
Thanks for the comment. I have to admit that my hands-on HPLC experience is pretty old at this point. While I still consider myself an analytical chemist, it's probably been at least 15 years since I've actually touched an HPLC. I did find the number/magnitude of peaks in the wash phase of the gradient surprising, but I suppose I didn't have reason to question it, as all of the peaks of interest come off well before the wash.Hi Wisco,
That is a very important distinction, and I'm sure many of us did not realize that was a blank/diluent injection. If this is a blank injection, there is certainly something on the column or in your injection matrix that is causing all of those peaks, and no, that amount of peaks is pretty irregular for blank/diluent injections, this should be a concern.
Yes, the method is fully validated at this point. May be worth a trial using a reference wavelength just to rule out RI/Temp effects, even if that change can't be easily implemented. If we find that it is a temperature effect, we could ask them to run on systems that have active temp control on the flow cell (I believe Agilent 1260 PDAs have this feature).Changes in the RI of the MP is a good reason to use a reference wavelength:
https://www.chromatographyonline.com/vi ... ction-hplc
https://www.agilent.com/library/support ... faq182.pdf
Is the method validated or could the use of a reference wavelength be assessed
Good point, but I think we're being pretty conservative with the cleanup portion of the gradient (6 min @ 100% MPB). If you look at the example in the original post, there's nothing eluting for the last two minutes of the cleanup phase (ie, from 24 - 26 min). My interpretation of that is that the wash is adequate, but I could be misreading that.If the "cleanout" portion of a gradient is executed too quickly, it is possible to retain some undesirable material on a column and have more of it elute during the following injection's cleanout.
Not sure that this is what's happening here, but there is a mechanism for what you describe.