The pH stability is important quality/reproducibility issue. pH inside pores of the basic portion of your "mixed" phase can be different from pH of your mobile phase depending on the buffer used, pH of the buffer, concentration of the buffer and buffering capacity. In ACN/water (pH 7) pH inside pores of silica with amino group on the surface is going to be well above 7 (closer to 10) which will cause this portion of ligand to fly off the surface and "travel" to the particle with acidic sites, where it will be trapped, thus changing selectivity of the column even if you use it within recommended pH range (pH 7). You need to stay in acidic buffered pH within recommended pH range and I would suggest to stay below 7 (at least pH 6.5). Unless this is a "hybrid" material like Waters and Phenomenex the hydrolysis of "amino" ligand in certain mobile phase (pH 6.5-7) is inevitable event, no matter how well material is “blendedâ€