Posted: Tue May 26, 2009 3:13 am
I'm almost about to scream "PROBLEM RESOLVED" (almost!)
After trying flow rate adjustments which made no difference (thanks for the tip anyway!), I stood around looking at the two instruments and tried to figure out what was the easiest piece to pull and swap. The electron multiplier is pretty easy to remove, so I shut down both instruments, pulled both EMs, and put the newer one in the older instrument.
Until now the older GCMS has given consistently non-linear curves, with reponse increasing by a factor of ~3 when concentration doubles. After swapping the EM's I just got this below from the older GCMS. These are reinjections from Friday's solutions (http://i43.tinypic.com/ridjx3.jpg), so it's not great, but it is straight!
I want to switch back to the old EM and rerun again to confirm, as well as go back to some of the other assays I've been trying to work on, before I rule this issue as completely resolved.

After trying flow rate adjustments which made no difference (thanks for the tip anyway!), I stood around looking at the two instruments and tried to figure out what was the easiest piece to pull and swap. The electron multiplier is pretty easy to remove, so I shut down both instruments, pulled both EMs, and put the newer one in the older instrument.
Until now the older GCMS has given consistently non-linear curves, with reponse increasing by a factor of ~3 when concentration doubles. After swapping the EM's I just got this below from the older GCMS. These are reinjections from Friday's solutions (http://i43.tinypic.com/ridjx3.jpg), so it's not great, but it is straight!
I want to switch back to the old EM and rerun again to confirm, as well as go back to some of the other assays I've been trying to work on, before I rule this issue as completely resolved.