Loss of resolution after replacing inlet glass liner

Discussions about GC and other "gas phase" separation techniques.

32 posts Page 3 of 3

Peter and AICMM,
You guys solved the problem. Thank you for your persistence. I have learned a lot from you guys, you have been wonderful teachers and I feel like i need to take you guys out for a beer or something!! I never realized would have realized that I must have been running split because of a leak, while the conditions were basically splitless.
The peak comes out with not tailing now.
Thank you
AD
DocHollywood

Thanks for the feedback. If you are ever in northern Botswana there is a bar just down the road !

Peter
Peter Apps