Let's review:
1. You inject 1 ul of a 1mg/l solution splitless, that gives 1 ng per analyte on column. This give speaks that are too small for your purpose, and much smaller than you can achive injecting the same solution onto a different make of instrument.
2. As well as the peaks being too small the background at the retention times of the peaks is too high. The way that the background declines, and the changes in how it declines look like the tail of a solvent peak.
3. The MS detector passes its detectability tests with a different analyte and solvent.
4. The MS and the inlet consumables have been cleaned and / or swapped repeatedly, with no improvement in performance.
Troubleshooting:
Can you run the same phthallate solution on another instrument, even an FID ?
If you run a 100 mg/l solution what do the peaks and solvent tail look like ?
Do the spectra of the peaks confirm that they are what they are supposed to be ?
Does the spectrum of the background confirm that it is toluene ?
If you run any other solution on the GC-MS; alkanes in hexane for example (but not the Varian test mix), are there the same problems ?.
Peter